[RASMB] concentration range for non-ideal behavior
Allen Minton
minton at helix.nih.gov
Fri Jan 23 09:00:14 PST 2015
John -
Nonideal behavior is due to nonspecific
interactions between solute molecules that may be
weak on a pairwise basis, but become significant
with increasing solute concentration. These
interactions may be either repulsive or
attractive and depend upon a variety of factors,
including pH, ionic strength, and temperature. I
am attaching a paper with a very simplified
analysis of the pH-dependent nonideal behavior of
bovine serum albumin. The bottom line is that
there can be no "general rule" for a
concentration below which nonideal behavior is
negligible. Under some conditions a solution may
behave almost ideally up to quite high
concentration, and under other conditions
nonideal behavior will kick in at low
concentrations. You can see this in Fig 1 of the
attached article. Depending upon pH and the
concentration and type of counterions, a large
polyanion like DNA can behave nonideally at concentrations below 0.1 mg/ml. .
Allen Minton
At 03:02 PM 1/22/2015, John Sumida wrote:
>Content-Type: multipart/alternative;
> boundary="----=_NextPart_000_007F_01D0363B.5BC26020"
>Content-Language: en-us
>
>Dear RASMB,
>
>David and all who have responded, thank you for
>your comments. I think in the aggregate of
>comments received I have my answer and more questions.
>
>In the range of concentrations measured 0.2
>4.9mgs/ml the slope of the experimental s with
>concentration is negative but only slightly so
>(slope=-0.033 s/c, where c is the mass
>concentration g/cc). Â The extrapolated s at
>0mgs/ml is 6.88, given the information received
>in these posts I think that at 4.2 mgs/ml I
>ought to expect a ~4.4% decrease in s due solely
>to excluded volume effects. If so the expected
>s-exp at 4.9mgs/ml ought to be 6.58. What is
>actually observed is a value of 6.72s. So I
>think what David and others have suggested is
>probably accurate, namely that self-association
>is offsetting the non-ideality here.Â
>
>This raises some questions perhaps that would be
>good to test. I canât shut off
>self-association but perhaps one could slow it
>down by reducing the temperature from 20C to 4C?
>I think excluded volume is only weakly dependent
>on temperature and self-association more strongly dependent on temperature?Â
>
>Interestingly, the sw vs conc relationship
>changes as a function of buffer, whereas the
>slope of relationship is only weakly negative in
>HEPES (0.01M HEPES, 150mM NaCl, pH7.5) in PBS
>(0.137M NaCl) it is strongly positive.
>
>Once again thank you all for your thoughts and suggestions.
>
>Best regards,
>John Sumida
>University of Washington
>
>From: RASMB [mailto:rasmb-bounces at list.rasmb.org] On Behalf Of David Hayes
>Sent: Wednesday, January 21, 2015 6:25 PM
>To: RASMB
>Subject: Re: [RASMB] concentration range for non-ideal behavior
>
>Hi John,
>
>The opposite of non-ideality is reality. All
>real molecules take up space, so all real
>molecules sediment with some excluded volume
>non-ideality. Charge is another source of
>non-ideality: the ideal model of a molecule,
>besides being a geometric point with no volume,
>also has no attractive or repulsive forces.
>The type of non-ideality that is not avoidable
>is excluded volume. So it is not exactly mg/ml
>that determines whether or not the non-ideality
>is significant, but the excluded volume. For
>most proteins it is reasonable to see little
>effect from concentration alone up to 1-2
>mg/ml. But, put a molecule in a salt free
>buffer where charge is not screened (Actin) or
>take a molecule that has a really big
>carbohydrate attached (Pegasys) and nonideality
>can be evident at 0.5 mg/ml and below.
>
>In sedimentation velocity, the other
>non-ideality is the back flow. When a real
>molecule moves down the cell, real buffer has to
>move up the cell to fill in the otherwise to be
>empty spot where the molecule was. This
>hydrodynamic non-ideality is also always there,
>but usually negligible below 1-2 mg/ml.
>
>Did you do a concentration series of your
>antibody? If so, it is expected that around 5
>mg/ml; you would have measurable concentration
>dependence of the sedimentation coefficient. If
>your peak moves the same S at 0.5 and 5 mg/ml I
>would suspect self association is adding S and
>offsetting the concentration dependence of
>S. Static light scattering is helpful to use as
>an orthogonal method when self association and
>non-ideality balance out in AUC-SV.
>
>David
>
>
>
>
>From: John Sumida
><<mailto:jpsumida at u.washington.edu>jpsumida at u.washington.edu>
>To: <mailto:rasmb at rasmb.org>rasmb at rasmb.org
>Sent: Tuesday, January 20, 2015 9:00 PM
>Subject: [RASMB] concentration range for non-ideal behavior
>
>Dear RASMB,
>
>My recollection is that concentrations in excess
>of 1.0 mg/ml should be avoided due to the
>possibility of non-ideal interactions that would
>complicate the analysis of SV and SE
>measurements. However in two recent cases (in
>one case a mAb was being studied, and in another
>a globular viral capsid protein), we exceeded
>this value by a factor of five and failed to
>observe any evidence of non-ideal behavior. In
>both cases, it was possible to fit both SV and
>SE data without assuming any non-ideal
>interactions when the experiment was performed in PBS or HEPES.
>
>I would be curious to hear if it is the
>experience of the community that non-ideality is
>commonly to be expected at
>concentrations >1.0mg/ml or whether the
>concentration cutoff for non-ideality is less of
>an actual cutoff being more a function of the
>particular properties of the molecule being studied.
>
>Best regards
>John Sumida
>University of Washington
>
>_______________________________________________
>RASMB mailing list
><mailto:RASMB at list.rasmb.org>RASMB at list.rasmb.org
>http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>
>_______________________________________________
>RASMB mailing list
>RASMB at list.rasmb.org
>http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20150123/17f79e65/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: BPC 95 crowding osm press.pdf
Type: application/pdf
Size: 536853 bytes
Desc: not available
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20150123/17f79e65/attachment-0002.pdf>
More information about the RASMB
mailing list