[RASMB] XLI absorbance problem

Andrew Leech andrew.leech at york.ac.uk
Thu Feb 26 09:19:25 PST 2015


Hello all,

I should learn to do the experiment before making pronouncements!
The counterbalance radial intensity scan on my XL/I gives about
150 counts where the light is coming through at the edges, so not
too different from Igor's system.

Andrew

On 24/02/2015 10:29, Andrew Leech wrote:
> Hi Igor,
>
> Those numbers (80 counts) look really low. I only recall seeing
> such small numbers when we had a photomultiplier problem, but your
> other scans rule that out. I think the light is not going through
> the counterbalance holes and you are seeing stray light.
>
> All the best,
>
> Andrew
>
>
> On 24/02/2015 08:01, Igor Perevyazko wrote:
>> Dear Titus,
>>
>> Please find attached the radial intensity scan of the counter balance.
>>
>> Best regards
>> Igor
>>
>> On Sun, Feb 22, 2015 at 9:09 PM, Titus M. Franzmann
>> <titus.franzmann at outlook.com <mailto:titus.franzmann at outlook.com>> wrote:
>>
>>     Can you send a radial scan of the counter balance.
>>
>>     Sent from my Windows Phone
>>
>> ------------------------------------------------------------------------
>>     Von: Igor Perevyazko <mailto:i.perevyazko at gmail.com>
>>     Gesendet: ‎22.‎02.‎2015 18:06
>>     An: jphilo at mailway.com <mailto:jphilo at mailway.com>; rasmb at rasmb.org
>>     <mailto:rasmb at rasmb.org>
>>     Betreff: Re: [RASMB] XLI absorbance problem
>>
>>     Dear John,
>>
>>     Thank you very much for your reply.
>>
>>     You wrote:
>>      > Is there any possibility the instrument is somehow reporting data
>>     for the wrong cell? You have told the GUI that you are using a
>>     4-hole rotor, not an 8-hole, correct? What was in the other cell
>>     positions when you recorded these benzene data?
>>
>>     Yes,I was using 4-hole rotor and GUI was checked for it as well.
>>     The last two experiments (* Radial intensity scan of KNO3 (0.7
>>     g/dl), with 0.6OD at 302 <mailto:0.6OD at 302> nm ** Radial intensity scan
>>     of Pyrene (2.5 microM), with ~1.2OD at 3 <mailto:0.6OD at 302>38 nm) were
>>     conducted separately, with only one cell loaded. I don't think that
>>     the instrument reporting data for the wrong cell. For some reason
>>     the photomultiplier does not see the difference in
>>     intensities/absorbance for samples which definitely absorb, unless a
>>     real big concentration(that is how we can also proof  that the
>>     instrument doesn't mismatched the position of a scanning cell) is
>> used.
>>     As I mention previously, the absorbance of samples was checked on
>>     regular Spectrofotometer and on working XLI_2. We have as well
>>     checked the monochromator of our XLI_1 on the working XLI_2 (please,
>>     see the initial message). This makes us think, that monochromator
>> is ok.
>>
>>     Best regards
>>     Igor
>>
>>
>>
>>     On Sun, Feb 22, 2015 at 2:09 AM, John Philo <jphilo at mailway.com
>>     <mailto:jphilo at mailway.com>> wrote:
>>
>>         Igor, I agree with Tom that possibly the wavelength is very
>>         different from what is reported. For that reason it would
>>         perhaps be better to test with something having very broad
>>         absorbance rather than sharp peaks. On the other hand the
>>         wavelength intensity scans you sent earlier seem to indicate the
>>         wavelength is getting selected (i.e. the grating has not fallen
>>         out) and not grossly out of calibration.____
>>
>>         __ __
>>
>>         Is there any possibility the instrument is somehow reporting
>>         data for the wrong cell? You have told the GUI that you are
>>         using a 4-hole rotor, not an 8-hole, correct? What was in the
>>         other cell positions when you recorded these benzene data?____
>>
>>         __ __
>>
>>         John____
>>
>>         __ __
>>
>>         *From:*RASMB [mailto:rasmb-bounces at list.rasmb.org
>>         <mailto:rasmb-bounces at list.rasmb.org>] *On Behalf Of *Igor
>>         Perevyazko
>>         *Sent:* Saturday, February 21, 2015 11:24 AM
>>         *To:* Laue, Thomas
>>
>>
>>         *Cc:* RASMB List
>>         *Subject:* Re: [RASMB] XLI absorbance problem____
>>
>>         __ __
>>
>>         Dear Thomas,____
>>
>>         Thank you very much for your reply.____
>>
>>         Well, the same intensities points out the problem we are dealing
>>         with (nonetheless, at much higher concentration (~ 30 mg/ml for
>>         proteins, for example) we do observe some level of optical
>>         densities ~0.1-0.2 OD). As I mentioned previously we have
>>         checked our monochromator on another XLI and it was working
>>         absolutely fine (please see the first email in the conversion
>>         list, there is also information about the lamp intensity and its
>>         comparison with the intensities of dye solution). The mismatched
>>         position of menisci, you have mentioned, I believe  is simply
>>         related to a slightly different filling volumes. ____
>>
>>         The long -pass filter was in 0 zero position. ____
>>
>>         __ __
>>
>>         Best regards____
>>
>>         Igor____
>>
>>         __ __
>>
>>         On Sat, Feb 21, 2015 at 9:18 PM, Laue, Thomas <Tom.Laue at unh.edu
>>         <mailto:Tom.Laue at unh.edu>> wrote:____
>>
>>             Hi-
>>             The sample and reference intensities are nearly identical.
>>             There is no way they are 1.2 and 0.6 OD.
>>             Is there any possibility that your monochromator is
>>             malfunctioning? If the actual wavelength were significantly
>>             different from what is being reported, you would get the
>>             sort of scans you are seeing. The menisci are mismatched,
>>             demonstrating that there is not a leak across the septum, so
>>             a wrong wavelength makes some sense.
>>             One last possibility- is the 400 nm long-pass filter in
>>             (lever in the horizontal position)? That would block the
>>             signal at the shorter wavelengths. The instrument would
>>             respond by setting the gain as high as it could, leading to
>>             a noisier signal (all dark current).
>>             Tom____
>>
>>
>> ------------------------------------------------------------------------
>>
>>             *From:*RASMB [rasmb-bounces at list.rasmb.org
>>             <mailto:rasmb-bounces at list.rasmb.org>] on behalf of Igor
>>             Perevyazko [i.perevyazko at gmail.com
>>             <mailto:i.perevyazko at gmail.com>]
>>             *Sent:* Saturday, February 21, 2015 11:49 AM
>>             *To:* Luitgard.Nagel-Steger at uni-duesseldorf.de
>>             <mailto:Luitgard.Nagel-Steger at uni-duesseldorf.de>;
>>             jphilo at mailway.com <mailto:jphilo at mailway.com>
>>             *Cc:* RASMB List
>>             *Subject:* Re: [RASMB] XLI absorbance problem____
>>
>>             Dear John, Luitgard____
>>
>>             Attached you can find radial intensity scans with an optical
>>             density of ~0.6 OD (KNO3, с = 0.07 dl/g, l = 302 nm) and
>>             ~1.2 OD (pyrene dye, 2.5micrM, l = 338 nm). I am using
>>             4-hole rotor. ____
>>
>>             Regarding Luitgards points: Concerning benzene, this is
>>             correct, but for our system,  previous experiments with such
>>             solvent /polymer combination resulted in a satisfactory
>>             optical densities. The large concentration of BSA in this
>>             case was chosen, as a model/test system to represent the
>>             problem - that even at such a high BSA content the optical
>>             density is at zero level.____
>>
>>             Looking forward to any of your suggestions.____
>>
>>             __ __
>>
>>             Best regards____
>>
>>             Igor____
>>
>>             __ __
>>
>>             On Fri, Feb 20, 2015 at 9:12 PM, Luitgard Nagel-Steger
>>             <Luitgard.Nagel-Steger at uni-duesseldorf.de
>>             <mailto:Luitgard.Nagel-Steger at uni-duesseldorf.de>> wrote:____
>>
>>                 Dear Igor,
>>
>>                 I am wondering whether benzene is a good choice of a
>>                 solvent if your pyrene dye is expected to absorb in the
>>                 range between 250 and 350 nm (shown in the attached
>>                 graph 1), since it absorbs itself rather strongly in the
>>                 UV range.
>>
>>                 Also 4 mg/ml of BSA are a pretty large amount of protein
>>                 for the optics in the AUC. Assuming you are using
>>                 standard cells with 1.2 cm optical path length this
>>                 should give an absorbance at 280 nm of 4 OD. The upper
>>                 limit is at about 1 OD.
>>
>>                 Best regards,
>>
>>                 Luitgard
>>
>>                 Am 20.02.2015 um 14:20 schrieb Igor Perevyazko:____
>>
>>                     Dear RASMB members,
>>
>>                     We are having a problem with absorbance optics on
>>                     XLI centrifuge(XLI_1
>>                     on the images attached). Hopefully you can suggest
>>                     something that will
>>                     help. The problem is that optical density of any
>>                     samples (which
>>                     definitely absorb, including for example BSA (c =
>>                     4mg/mL) or pyrene dye
>>                     (c = 10µM )) recorded by the AUC is within the zero
>>                     level.
>>                     Nevertheless, at much higher concentrations
>>                     (approximately 10 -20 times
>>                     higher) the optical density reaches some level of
>>                     absorbance (~0.1-0.2
>>                     OD). The same solutions checked by normal UV-Vis
>>                     spectrophotometer and
>>                     on another XLI centrifuge (XLI_2), have adequate
>>                     values of the
>>                     absorbance level. We have as well checked our
>>                     monochromator   on  XLI_2
>>                     and it was working ok. Recorded lamp intensities
>>                     (air to air) at 230 nm
>>                     and 530 nm are at the normal level. If we perform
>>                     the absorbance scan of
>>                     a sample (for example Pyrene dye at c = 10µM)  in
>>                     the intensity mode, we
>>                     received the same level of intensities for both
>>                     reference and sample
>>                     chamber! The corresponding comparison of the
>>                     intensities is shown on the
>>                     image attached. Any of your suggestions about the
>>                     possible problem
>>                     and/or the solution will be highly appreciated.
>>
>>
>>                     Best regards
>>
>>                     Igor Perevyazko
>>
>>
>>                     PhD Igor Perevyazko
>>                     Physics Department, Polymer Physics
>>                     St.Petersburg State University
>>                     Ul. Ulyanovskaya 1, Petrodvorets
>>                     St.Petersburg, Russia, 198504
>>                     Tel.: (812)4284382
>>                     Fax.: (812)4287240
>>
>>
>>                     ____
>>
>>                     _______________________________________________
>>                     RASMB mailing list
>>                     RASMB at list.rasmb.org <mailto:RASMB at list.rasmb.org>
>>
>> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org____
>>
>>                 _______________________________________________
>>                 RASMB mailing list
>>                 RASMB at list.rasmb.org <mailto:RASMB at list.rasmb.org>
>>                 http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org____
>>
>>             __ __
>>
>>         __ __
>>
>>
>>
>>
>>
>> _______________________________________________
>> RASMB mailing list
>> RASMB at list.rasmb.org
>> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>>
>

-- 
Dr Andrew Leech                   *  Laboratory Head
Technology Facility               *  Molecular Interactions Laboratory
Department of Biology (Area 15)   *  Tel   : +44 (0)1904 328723
University of York                *  Fax   : +44 (0)1904 328804
Wentworth Way,  York  YO10 5DD    *  Email : andrew.leech at york.ac.uk
EMAIL DISCLAIMER: http://www.york.ac.uk/docs/disclaimer/email.htm