Hello Andrew,


Thanks for the tip, it make sense. I didn't try that yet, but will do it
tomorrow.

Best regards
Igor

On Sun, Feb 22, 2015 at 8:20 PM, Andrew Leech <andrew.leech at york.ac.uk>
wrote:

> Hi Igor,
>
> Have you tried loading the cell upside down (or backwards), to put the
> sample in the reference side and vice versa? Alternatively load the
> two sectors with grossly different volumes (only spin at 3000 rpm to
> scan). I think you are seeing the reference sector twice somehow.
>
> All the best,
>
> Andrew
>
> On 21/02/2015 22:09, John Philo wrote:
>
>> Igor, I agree with Tom that possibly the wavelength is very different
>> from what is reported. For that reason it would perhaps be better to
>> test with something having very broad absorbance rather than sharp
>> peaks. On the other hand the wavelength intensity scans you sent earlier
>> seem to indicate the wavelength is getting selected (i.e. the grating
>> has not fallen out) and not grossly out of calibration.
>>
>> Is there any possibility the instrument is somehow reporting data for
>> the wrong cell? You have told the GUI that you are using a 4-hole rotor,
>> not an 8-hole, correct? What was in the other cell positions when you
>> recorded these benzene data?
>>
>> John
>>
>> *From:*RASMB [mailto:rasmb-bounces at list.rasmb.org] *On Behalf Of *Igor
>> Perevyazko
>> *Sent:* Saturday, February 21, 2015 11:24 AM
>> *To:* Laue, Thomas
>> *Cc:* RASMB List
>> *Subject:* Re: [RASMB] XLI absorbance problem
>>
>> Dear Thomas,
>>
>> Thank you very much for your reply.
>>
>> Well, the same intensities points out the problem we are dealing with
>> (nonetheless, at much higher concentration (~ 30 mg/ml for proteins, for
>> example) we do observe some level of optical densities ~0.1-0.2 OD). As
>> I mentioned previously we have checked our monochromator on another XLI
>> and it was working absolutely fine (please see the first email in the
>> conversion list, there is also information about the lamp intensity and
>> its comparison with the intensities of dye solution). The mismatched
>> position of menisci, you have mentioned, I believe  is simply related to
>> a slightly different filling volumes.
>>
>> The long -pass filter was in 0 zero position.
>>
>> Best regards
>>
>> Igor
>>
>> On Sat, Feb 21, 2015 at 9:18 PM, Laue, Thomas <Tom.Laue at unh.edu
>> <mailto:Tom.Laue at unh.edu>> wrote:
>>
>>     Hi-
>>     The sample and reference intensities are nearly identical. There is
>>     no way they are 1.2 and 0.6 OD.
>>     Is there any possibility that your monochromator is malfunctioning?
>>     If the actual wavelength were significantly different from what is
>>     being reported, you would get the sort of scans you are seeing. The
>>     menisci are mismatched, demonstrating that there is not a leak
>>     across the septum, so a wrong wavelength makes some sense.
>>     One last possibility- is the 400 nm long-pass filter in (lever in
>>     the horizontal position)? That would block the signal at the shorter
>>     wavelengths. The instrument would respond by setting the gain as
>>     high as it could, leading to a noisier signal (all dark current).
>>     Tom
>>
>>     ------------------------------------------------------------
>> ------------
>>
>>     *From:*RASMB [rasmb-bounces at list.rasmb.org
>>     <mailto:rasmb-bounces at list.rasmb.org>] on behalf of Igor Perevyazko
>>     [i.perevyazko at gmail.com <mailto:i.perevyazko at gmail.com>]
>>     *Sent:* Saturday, February 21, 2015 11:49 AM
>>     *To:* Luitgard.Nagel-Steger at uni-duesseldorf.de
>>     <mailto:Luitgard.Nagel-Steger at uni-duesseldorf.de>;
>>     jphilo at mailway.com <mailto:jphilo at mailway.com>
>>     *Cc:* RASMB List
>>     *Subject:* Re: [RASMB] XLI absorbance problem
>>
>>     Dear John, Luitgard
>>
>>     Attached you can find radial intensity scans with an optical density
>>     of ~0.6 OD (KNO3, с = 0.07 dl/g, l = 302 nm) and ~1.2 OD (pyrene
>>     dye, 2.5micrM, l = 338 nm). I am using 4-hole rotor.
>>
>>     Regarding Luitgards points: Concerning benzene, this is correct, but
>>     for our system,  previous experiments with such solvent /polymer
>>     combination resulted in a satisfactory  optical densities. The large
>>     concentration of BSA in this case was chosen, as a model/test system
>>     to represent the problem - that even at such a high BSA content the
>>     optical density is at zero level.
>>
>>     Looking forward to any of your suggestions.
>>
>>     Best regards
>>
>>     Igor
>>
>>     On Fri, Feb 20, 2015 at 9:12 PM, Luitgard Nagel-Steger
>>     <Luitgard.Nagel-Steger at uni-duesseldorf.de
>>     <mailto:Luitgard.Nagel-Steger at uni-duesseldorf.de>> wrote:
>>
>>         Dear Igor,
>>
>>         I am wondering whether benzene is a good choice of a solvent if
>>         your pyrene dye is expected to absorb in the range between 250
>>         and 350 nm (shown in the attached graph 1), since it absorbs
>>         itself rather strongly in the UV range.
>>
>>         Also 4 mg/ml of BSA are a pretty large amount of protein for the
>>         optics in the AUC. Assuming you are using standard cells with
>>         1.2 cm optical path length this should give an absorbance at 280
>>         nm of 4 OD. The upper limit is at about 1 OD.
>>
>>         Best regards,
>>
>>         Luitgard
>>
>>         Am 20.02.2015 um 14:20 schrieb Igor Perevyazko:
>>
>>             Dear RASMB members,
>>
>>             We are having a problem with absorbance optics on XLI
>>             centrifuge(XLI_1
>>             on the images attached). Hopefully you can suggest something
>>             that will
>>             help. The problem is that optical density of any samples
>> (which
>>             definitely absorb, including for example BSA (c = 4mg/mL) or
>>             pyrene dye
>>             (c = 10µM )) recorded by the AUC is within the zero  level.
>>             Nevertheless, at much higher concentrations (approximately
>>             10 -20 times
>>             higher) the optical density reaches some level of absorbance
>>             (~0.1-0.2
>>             OD). The same solutions checked by normal UV-Vis
>>             spectrophotometer and
>>             on another XLI centrifuge (XLI_2), have adequate values of the
>>             absorbance level. We have as well checked our monochromator
>>               on  XLI_2
>>             and it was working ok. Recorded lamp intensities (air to
>>             air) at 230 nm
>>             and 530 nm are at the normal level. If we perform the
>>             absorbance scan of
>>             a sample (for example Pyrene dye at c = 10µM)  in the
>>             intensity mode, we
>>             received the same level of intensities for both reference
>>             and sample
>>             chamber! The corresponding comparison of the intensities is
>>             shown on the
>>             image attached. Any of your suggestions about the possible
>>             problem
>>             and/or the solution will be highly appreciated.
>>
>>
>>             Best regards
>>
>>             Igor Perevyazko
>>
>>
>>             PhD Igor Perevyazko
>>             Physics Department, Polymer Physics
>>             St.Petersburg State University
>>             Ul. Ulyanovskaya 1, Petrodvorets
>>             St.Petersburg, Russia, 198504
>>             Tel.: (812)4284382
>>             Fax.: (812)4287240
>>
>>
>>             _______________________________________________
>>             RASMB mailing list
>>             RASMB at list.rasmb.org <mailto:RASMB at list.rasmb.org>
>>             http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>>
>>         _______________________________________________
>>         RASMB mailing list
>>         RASMB at list.rasmb.org <mailto:RASMB at list.rasmb.org>
>>         http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>>
>>
>>
>> _______________________________________________
>> RASMB mailing list
>> RASMB at list.rasmb.org
>> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>>
>>
> --
> Dr Andrew Leech                   *  Laboratory Head
> Technology Facility               *  Molecular Interactions Laboratory
> Department of Biology (Area 15)   *  Tel   : +44 (0)1904 328723
> University of York                *  Fax   : +44 (0)1904 328804
> Wentworth Way,  York  YO10 5DD    *  Email : andrew.leech at york.ac.uk
> EMAIL DISCLAIMER: http://www.york.ac.uk/docs/disclaimer/email.htm
>