Hi John,
The accuracy of CFCA method depends on the actual protein and the assay setup. From our experience it is anywhere between 10 to 20%. One normally runs at nM concentration of the samples and assumes that the mass transport is controlled by the diffusion of the monomeric species. You need to know D - measure, calculate, find in literature etc. Samples have to be homogeneous for the best performance of the method. Variability of glycosylation level will add to the inaccuracy .
Best regards,
Natalia
From: John Sumida [mailto:jpsumida at u.washington.edu]
Sent: den 14 januari 2014 18:39
To: 'Mattia Rocco'; Markova, Natalia (GE Healthcare); rasmb at rasmb.org
Subject: RE: [RASMB] SPR query about CFCA applications
Natalia and Mattia,
Thank you for your responses. I gather that in general these values are not measured for the CFCA analysis and that perhaps there is not sufficient reason to do so given the precision of the application? I wonder how variably glycosylated proteins would affect the accuracy of this technique?
Thank you both again for taking the time to respond.
Best regards,
John Sumida, Ph.D.
Analytical Biopharmacy Core Facility<http://depts.washington.edu/cidb4bio/index.shtml>
University of Washington
Molecular Engineering & Sciences Institute, G22<http://www.moles.washington.edu/>
3946 West Stevens Way NE
Seattle WA 98195-1651
From: Mattia Rocco [mailto:mattia.rocco at hsanmartino.it]
Sent: Tuesday, January 14, 2014 1:17 AM
To: Markova, Natalia (GE Healthcare); John Sumida; rasmb at rasmb.org<mailto:rasmb at rasmb.org>
Subject: Re: [RASMB] SPR query about CFCA applications
The f/f0 Natalia mentions is for almost spherical shape.... if you are dealing with an antibody, then the 1.5 value might be more appropriate (but I haven't done any calculations on it... ;-) )
Best - Mattia
At 08:13 AM 1/14/2014 +0000, Markova, Natalia (GE Healthcare) wrote:
Content-Language: en-US
Content-Type: multipart/alternative;
boundary="_000_41C5780FD432C74DB68C588BE66E34F614EF6B76BUDURBPA08e2kad_"
Hi John,
We use as default f/f0 factor for globular proteins, which is 1.2 (not 1.5). However, user can freely enter any value for this factor in Web diffusion coefficient calculator (https://www.biacore.com/lifesciences/Application_Support/online_support/Diffusion_Coefficient_Calculator/index.html?section=lifesciences&realsection=lifesciences)
You need to be registered on the Biacore website to access it.
Best regards,
Natalia
Natalia Markova (PhD)
GE Healthcare
R&D Life Sciences
Rapsgatan 23
751 84 Uppsala
Sweden
M +46 727 33 58 80
E natalia.markova at ge.com<mailto:natalia.markova at ge.com>
www.biacore.com<http://www.biacore.com/>
www.microcal.com<https://exchange1.ad.slu.se/owa/redir.aspx?C=61606803db87492ba25280057e5d208e&URL=http%3a%2f%2fwww.microcal.com%2f>
Disclaimer
The information contained in this e-mail is intended solely for the use of the addressee and contains information that is confidential and may be legally privileged. Access to this e-mail by anyone else is unauthorised. If you are not the intended recipient, your use, retention, copy, any disclosure, dissemination or any action taken or omitted to be taken in reliance of this communication is strictly prohibited and may be unlawful. If you have received this communication in error, please notify us immediately by phone. Thank you!
From: rasmb-bounces at list.rasmb.org<mailto:rasmb-bounces at list.rasmb.org> [mailto:rasmb-bounces at list.rasmb.org] On Behalf Of John Sumida
Sent: den 14 januari 2014 00:20
To: rasmb at rasmb.org<mailto:rasmb at rasmb.org>
Subject: [RASMB] SPR query about CFCA applications
Dear RASMB,
This question is in regards to an SPR application. We are currently investigating the use calibration free concentration analysis (CFCA) to monitor the stability of macromolecular preps over time in terms of their activity. Input parameters for this analysis are ff0, viscosity, MW and D and we are using AUC to better determine both. Current applications seem to assume ff0=1.5, which may not be correct for monoclonal preps based on my reading of the literature.
My query.
1. Is it currently the common practice to measure these quantities and if so what is the preferred method to do so?
Thank you all and Happy New Year,
John Sumida, Ph.D.
Analytical Biopharmacy Core Facility<http://depts.washington.edu/cidb4bio/index.shtml>
University of Washington
Molecular Engineering & Sciences Institute, G22<http://www.moles.washington.edu/>
3946 West Stevens Way NE
Seattle WA 98195-1653
_______________________________________________
RASMB mailing list
RASMB at list.rasmb.org<mailto:RASMB at list.rasmb.org>
http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
ATTENZIONE.
Il presente messaggio ed i suoi allegati devono intendersi ad uso esclusivo dei suoi destinatari e sono confidenziali. Persone diverse dal destinatario, anche ai sensi del D.Lgs. n. 196/2003 , non debbono prendere visione dei contenuti, copiare od inoltrare l'e-mail a terzi. Se ricevete questo messaggio per errore, Vi preghiamo di cancellarlo, di distruggerne ogni copia e di informarci immediatamente. La mail non è un protocollo sicuro, pertanto IST declina ogni responsabilità in caso di intercettazione o modifiche del presente messaggio.
WARNING.
This message and any attachments is intended solely for the use of the intended
addressees and is confidential. Any unauthorized review, use, disclosure or distribution
is prohibited. If you receive this message in error, please delete it, destroy all copies
and immediately notify us. Mail is not a secure protocol, therefore IST will not be liable
for interception or amendment of this message.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20140119/f87982f4/attachment-0001.html>
More information about the RASMB
mailing list