Dear Ronald Toth, dear all,
to your question, there is a very simple answer: first of all, if the particle you want to sediment is equal to one in
water, it will not sediment but distribute all over the cell, so you can't identify any gradient near the cell meniscus or bottom.
Usually lipid particles (if they are homogen in density but usually they are not ) at the density range of ro± ...1
-0.1 or more then +1.0 in buffers. To identify them you should use schlieren optics since usually they do not absorb!
if they are not coloured.
To determine the true density of the lipid you have to add to your solvent/buffer a certain percent of
D2O to the solution. You have to perform several sed. velocity runs at different D2O concentrations until you can detect a band/ peak that has a negative sed.value,= that floatats. If you add too much D2O it will exceed the mid of the radial direction of the cell and be close to the meniscus.
In any case do not use for your sed. runs the density that you have determined with the advised upper predetermined runs . To calculate a true sedimentation you need to calculate the density of the particle.
For Mw =sed. equilibrium you need to do the same proceedure…yours …ariel. A retired UC user