[RASMB] XL-I Diagnostic 61 Error: Drive Malfunction
Toth IV, Ronald T
Toth.Ronald at ku.edu
Fri Mar 21 13:29:25 PDT 2014
Hello members!
I was wondering if any of you have experience with the Diag 61: Drive malfunction error on the XL-I. Beckman says it will take a month to get an engineer out and I'm wondering if it's something I could resolve myself, as we have samples that need to be run.
The error displayed on the machine is as follows:
Diag 61: Drive malfunction. Door will remain closed for one hour
>From what I've gathered so far, this may be a failure of the machine to sense the rotor, and it wants to wait an hour because of the danger that the rotor may still be coasting in the vacuum. However, even after several hours the error message will still not clear and allow the door to be opened.
After turning off the machine and releasing the vacuum and the door manually, I have seen that the rotor is not still spinning, however even after returning the vacuum screw to its original position and turning the machine back on with the rotor in the chamber, the error will not clear.
The relevant log files are as follows, thank you all in advance for any advice or tips you may have!
005637 SCAN Requesting new rotor scan
005638 SCAN scan 657 started
005638 FILE Updating cell1abs.log
090321 COM Could not write to com port.
090321 COM Break condition
090548 SCAN scan 657 stopped
- Ronald Toth
Post Doctoral Researcher
Macromolecule and Vaccine Stabilization Center
University of Kansas
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20140321/b7318f28/attachment.htm>
More information about the RASMB
mailing list