Richard,
in principle, a differential viscometer can yield not only relative, but
also absolute dynamic viscosities. Both p_i, the inlet pressure, and
delta p, the pressure difference between the solution and the solvent
branch, are absolute values - the question is, does the instrument
software give you access to these raw data. The maths I used, however,
only worked out for symmetric branches, and the instrument we used at
that time, a WGE eta-1002, had assymmetrical branches. I never got to
the point of solving the more complicated case. Eventually, the
viscometer broke down, and I abandoned the project.
However, a differential viscometer intended for use as an SEC detector
is very sensitive, and our measurements gave good data also for diluted
solutions. Consider that we had a constant eluent flow and injected as
little as 100 µl into that flow. So the signal was actually for a
solution yet more diluted than the bulk solution. The intriguing point
about this project was that if you couple a second detector into the
flow, like a UV, for concentration of the solute, you get two sets of
data: absolute viscosities and absolute concentrations for every data
point of the elugram. This is equivalent to a number of independent
measurements at varied concentration. Thus, it should in principle be
possible to calculate even an intrinsic viscosity from a single shot.
The drawback is that the eluent flow is split into the two branches,
where a reservoir on the solvent branch, filled with solvent at the
beginning of the experiment, gradually fills with solution containing
the solute. You have only a couple of shots until the solvent branch is
contaminated with solute, and the system needs to be purged.
Though I never brought this to an end, it still appears a promising
option that should work out - in principle.
Kind regards,
Kristian
Am 07.06.2013 05:53, schrieb Richard Kingston:
>
>
> Thanks for the many informative replies, both on- and off- list.
>
> The consensus seems to be that differential (pressure-imbalance)
> viscometers like the Viscostar are well suited to the measurement of
> protein intrinsic viscosities, and there are several in commercial
> production. However these instruments, by design, measure only
> relative viscosities
>
> Reframing my original question, I was really wondering if it were
> possible to find an instrument which could measure absolute solution
> viscosities with sufficient accuracy to allow protein characterisation
> (and could therefore do double duty, being able to measure solution
> viscosity for other biophysical applications). However for the
> reasons outlined by Arthur, and as attested below by Dave, it might be
> difficult to achieve good results this way.
>
> Best,
>
> Richard
>
--
Nanolytics
Gesellschaft fuer Kolloidanalytik mbH
Dr. Kristian Schilling
Am Muehlenberg 11
D-14476 Potsdam
Tel: +49 331 5818360
Fax: +49 331 5818361
e-mail: schilling at nanolytics.de
Internet: www.nanolytics.de
_______________________________________________
Diese E-Mail kann vertrauliche und/oder rechtlich geschützte Informationen enthalten. Wenn Sie nicht der richtige Adressat sind oder diese E-Mail irrtümlich erhalten haben, informieren Sie bitte sofort den Absender und vernichten Sie diese Mail. Das unerlaubte Kopieren sowie die unbefugte Weitergabe dieser Mail ist nicht gestattet.
This e-mail may contain confidential and/or privileged information. If you are not the intended recipient (or have received this e-mail in error) please notify the sender immediately and destroy this e-mail. Any unauthorized copying, disclosure or distribution of the material in this e-mail is strictly forbidden.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20130607/0278f167/attachment-0002.html>
More information about the RASMB
mailing list