[RASMB] Time-stamp checks for SV data
Ghirlando, Rodolfo (NIH/NIDDK) [E]
RodolfoG at intra.niddk.nih.gov
Fri Jan 25 07:59:37 PST 2013
Hi Borries -
It would appear as though the values of w^2t in the file header correlate with the adjacent time stamp in seconds. It could be that the time stamp is obtained through w^2t.
Whatever it is, these times (be they t or w^2t) eventually lag date and time stamps on the XLA/I (as obtained from the cell log files) and Windows OS on the PC.
Thanks.
Kind regards, Rodolfo
-----Original Message-----
From: Walter Stafford [mailto:wstafford3 at walterstafford.com]
Sent: Friday, January 25, 2013 10:38 AM
To: Borries Demeler
Cc: Ghirlando, Rodolfo (NIH/NIDDK) [E]; rasmb at list.rasmb.org
Subject: Re: [RASMB] Time-stamp checks for SV data
Borries,
The time values in the file header are "elapsed time" from the the time the rotor starts turning and should not be used in analysis in any case. SEDANAL uses integral(w^2 dt), which is the correct value to use for all analyses.
So your question is important: Are those values reported correctly or not in version 6?
Walter
_________________________
Walter Stafford
wstafford3 at walterstafford.com
_________________________
Walter Stafford
wstafford3 at walterstafford.com
On Jan 25, 2013, at 10:13, Borries Demeler wrote:
> Rodolfo:
>
> are w^2t values reported in the file header not affected, or has that not been
> investigated? I do not have this version to test myself, but such errors would
> propagate into some analysis methods in UltraScan as well. As far as I know
> the w^2t values are reported directly from the DA system of the XLA, I am not
> certain where the time values originate. The w^2t values *may* be affected as well,
> and that should be looked at by someone with a Proteomelab ver.6.
>
> -borries
>
>
> On Fri, Jan 25, 2013 at 09:29:39AM -0500, Ghirlando, Rodolfo (NIH/NIDDK) [E] wrote:
>> Dear friends and colleagues,
>>
>> It has come to our (Joy Zhao, Peter Schuck, Grzegorz Piszczek, Chad Brautigam and myself) attention that sedimentation coefficients based on SV data collected using the ProteomeLab XLA/I version 6 acquisition software are up to 10% larger than expected (depending on the rotor speed).
>>
>> We have traced this to a smaller than expected elapsed time stamp in the file header. A correction for this has been implemented in SEDFIT based in part on the Windows OS file creation time stamp. This version has been released, as indicated in the forwarded email from the SEDFIT-L list below, and we will report more details on this shortly.
>>
>> In addition Beckman Coulter has been advised of this issue.
>>
>> Sincerely,
>>
>> Rodolfo Ghirlando
>>
>>
>> FORWARDED MESSAGE
>>
>>
>> Dear Colleagues,
>>
>> This is to let you know of a critical SEDFIT update with version 14.0c, which you can download from https://sedfitsedphat.nibib.nih.gov/software/Shared%20Documents/sedfit140c.zip
>>
>> It has a very important new feature for the analysis of sedimentation velocity: it automatically compares the elapsed times reported in the data file headers with the time differences obtained from the file creation timestamps generated by the Windows operating system on the PC.
>>
>> Why would we need this? In the course of a joint study between our lab and the labs of Rodolfo Ghirlando, Greg Piszczek and Chad Brautigam, we have recently discovered that data files can have elapsed time entries less than the differences from the operating system timestamp, by as much as 10%. Obviously, if the elapsed time entry is 10% too low then this leads to s-values that are 10% too high!
>>
>> We have double checked data from many different instruments, and correlated this discrepancy in elapsed times with the ProteomeLab data acquisition software version 6. No time differences, or only trivial ones, were observed for instruments running earlier versions of the data acquisition software. In some runs the overall difference amounts to hours at the end of the run, and I believe that the operating system timestamp is correct. We will distribute more detailed information shortly, but want to give you already a heads-up in the meantime with this note: If you are running version 6 you may want to examine the data files, and/or time the data acquisition process yourself, to verify.
>>
>> We have discussed this matter with our Beckman service engineer and it appears as though the company is now addressing this issue.
>>
>> The new SEDFIT release can fix this problem for the data analysis. First, it automatically checks and creates a report if a systematic and significant discrepancy exists between differences in the file time-stamp and differences in the header elapsed time. If there appears to be a problem (i.e. apparent dilation factor greater than a user-defined threshold, initially defaulted to 1.005 or 0.5%), SEDFIT will create a report and then offer to write data files (under different names, of course) with time entries based on a dilation factor extracted from the file timestamps. Subsequent analysis of these time-corrected data files should eliminate the problem, as it did perfectly for the data in our study. Such time-corrected scan files should also be the ones loaded into SEDPHAT, which will get its own update shortly.
>>
>> Please let me know if you have any questions, or if any problems occur with the new version.
>>
>> Best wishes and good luck,
>> Peter
>>
>>
>> Peter Schuck, PhD
>> Chief, Dynamics of Macromolecular Assembly Section Laboratory of Cellular Imaging and Macromolecular Biophysics National Institute of Biomedical Imaging and Bioengineering, NIH
>> 13 South Drive
>> Bldg 13, Rm 3N17
>> Bethesda, MD 20892
>> phone: (301) 435-1950
>> fax: (301) 480-1242
>> email: schuckp at mail.nih.gov
>>
>>
>
>> _______________________________________________
>> RASMB mailing list
>> RASMB at list.rasmb.org
>> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>
> _______________________________________________
> RASMB mailing list
> RASMB at list.rasmb.org
> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
More information about the RASMB
mailing list