[RASMB] s20w calculation

Lake Paul lpaul at purdue.edu
Fri Aug 9 09:48:10 PDT 2013


John,
I would calculate it manually:
Eqn is S20W = s[(1-v*rho20W)/(1-v*rhoexperimental)*[nexperimental/n20W)

Here is the link to that equation:
http://bitcwiki.sr.unh.edu/index.php/Converting_s_to_s-20-w   (SEDNTERP)

When in doubt, don't trust HAL! :)

Lake



On Thu, Aug 8, 2013 at 8:38 PM, John Sumida <jpsumida at u.washington.edu> wrote:
> Dear RASMB
>
>
>
> Wondering about the s020w calculation in SEDNTERP
>
>
>
> For one of our polymer micelles with which we’ve been working, a SEDANAL
> analysis returns an s20w value (not corrected for concentration) of 17.3.
>
>
>
> If I use equation 19 in cited in the SEDNTERP documentation, from the s20w
> value=17.3, I calculate s* = 16.4s.
>
> Phsyical parameters used:
>
> 1.     experiment temp=20oC,
>
> 2.     vbar=0.8312 (measured by densitometry at 20oC)
>
> 3.     density of solution 1.00506 gms/cm3 (measured)
>
> 4.     viscosity 0.010281 (measured)
>
> 5.     Assuming viscosity of water to be 0.01002, and the density to be
> 0.998234
>
>
>
> However if I use the Experiment page in the SEDNTERP and enter value of s*
> calculated above, then s20w returned is 27.8 which is much higher than the
> value returned from SEDANAL.  In fact in order to get SEDNTERP to return a
> value of 17.3s, I have to decrease S* to be 2.91s.
>
>
>
> In SEDFIT, if I input 16.4s, calculated from Equation 19 in the SEDNTERP
> documentation, and input the same physical parameter listed above, I get
> 17.4s in agreement with the SEDANAL result.
>
>
>
> I think in both SEDNTERP and SEDFIT, the same equation is used to correct s*
> to s20w, so I’m not clear what I’m missing.
>
>
>
> Any suggestions and comments are welcome and thank you, in advance, for
> taking the time to consider my question.
>
>
>
> Best regards,
>
>
>
> John Sumida
>
> Analytical Biopharmacy Core Facility
>
> University of Washington
>
> Molecular Engineering & Sciences Institute, G22
>
> 3946 West Stevens Way NE
>
> Seattle WA 98195-1653
>
>
>
>
> _______________________________________________
> RASMB mailing list
> RASMB at list.rasmb.org
> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
>



-- 
Lake Paul, PhD
Facility Director

Biophysical Analysis Lab
Purdue Proteomics Facility

Bindley Bioscience Center
Purdue University
1203 West State Street,
West Lafayette, Indiana, 47095
765.494.4960



More information about the RASMB mailing list