[RASMB] Viscometer

Richard Kingston rl.kingston at auckland.ac.nz
Thu Jun 6 20:53:27 PDT 2013



Thanks for the many informative replies, both on- and off- list.

The consensus seems to be that differential (pressure-imbalance) viscometers like the Viscostar are well suited to the measurement of protein intrinsic viscosities, and there are several in commercial production. However these instruments, by design, measure only relative viscosities 

Reframing my original question, I was really wondering if it were possible to find an instrument which could measure absolute solution viscosities with sufficient accuracy to allow protein characterisation (and could therefore do double duty, being able to measure solution viscosity for other biophysical applications).  However for the reasons outlined by Arthur, and as attested below by Dave, it might be difficult to achieve good results this way.

Best,

Richard

On 7/06/2013, at 5:56 AM, David Scott <David.Scott at nottingham.ac.uk> wrote:

> Hello All in RASMB land,
> 
> We have both the viscostar and the Anton Paar viscometer in the NCMH, and have used both to try and attain viscosity increments. In practice, only the viscostar actually works at reasonable concentrations (c.a. 1 mg/ml) for proteins. We've used this to look at conformations of intrinsically disordered proteins:
> 
> Rajeskar, K.V., Muntaha, S.T., Tame, J.R.H., Wharton, C.M., Thomas, C.M., White, S.A., Hyde, E.I. and Scott, D.J. (2010). Order and disorder in the domain organisation of the plasmid partition protein KorB. J. Biol. Chem. 285 15440-15449
> 
> It gives very nice results, especially coupled to SEC-MALLS. I would also add to Mattia's point that using absorbance to measure your protein concentration is so much more accurate than refractive index increments. 
> 
> The Anton Paar system gives very good (3 d.p.) precision measurements of solution viscosities which are essential for accurate corrections of sedimentation velocity data, but it's use beyond this is limited, in my experience. I know others in the NCMH have tried to push it further with carbohydrate systems, but they tend to be more abundant in yields than proteins.
> 
> The only downside to the viscostar is a tendency to blockages, so needs to be used on a fairly constant basis, and washed through regularly. However, despite that caveat, I have been impressed with it (when it is working properly...).
> 
> Best
> 
> Dave Scott.
> 
> 
> 
> 
> 
> 
> 
> On 6 Jun 2013, at 17:22, Razinkov, Vladimir wrote:
> 
>> Dear all,
>>  
>> Another rather simple and high throughput  method to measure viscosity of proteins at low volumes (~ 35 uL with 384-well plate) using DLS plate reader.
>>  
>> Anal Biochem. 2010 Apr 1;399(1):141-3. doi: 10.1016/j.ab.2009.12.003. Epub 2009 Dec 6.
>> High-throughput dynamic light scattering method for measuring viscosity of concentrated protein solutions.
>> He F, Becker GW, Litowski JR, Narhi LO, Brems DN, Razinkov VI.
>> Source
>> 
>> Formulation and Analytical Resources, Amgen Inc., Seattle, WA 98119, USA.
>> 
>> Abstract
>> We propose a new method to measure the viscosity of concentrated protein solutions in a high-throughput format. This method measures the apparent hydrodynamic radius of polystyrene beads with known sizes using a dynamic light scattering (DLS) system with a microplate reader. Glycerol solution viscosities obtained by the DLS method were in good agreement with those reported in the literature. Viscosity of the solutions of two monoclonal antibody molecules was acquired using both DLS and cone-and-plate techniques, and the results were comparable. The DLS method described here has the potential to be used in many aspects of protein characterization
>> 
>>  
>>  
>> Regards
>>  
>> Vladimir Razinkov
>>  
>> From: rasmb-bounces at list.rasmb.org [mailto:rasmb-bounces at list.rasmb.org] On Behalf Of Richard Kingston
>> Sent: Wednesday, June 05, 2013 9:15 PM
>> To: rasmb at rasmb.org
>> Subject: [RASMB] Viscometer
>>  
>>  
>> Dear all,
>>  
>> I've become interested in measuring the viscosity of protein solutions at varying temperature,  and wondered if anyone on the list could advise on the best way to do this using commercially available instrumentation. While there are many  ways to measure viscosity, for protein applications minimizing the required sample volume is pretty critical. I note that  Anton Paar sell an instrument that measures both dynamic viscosity and density  (SVM 3000 Stabinger viscometer). If you skip the density measurement, the sample requirements look almost "manageable". Cambridge also sell a micro-sample viscometer (Viscolab 5000) which uses microlitre volumes. 
>>  
>> I'm having difficulty tracking down people that might have used these or similar instruments, and could comment on the practicalities for protein work. If anyone can offer advice, either on- or off-list, it would be appreciated.
>>  
>> Many thanks,
>>  
>> Richard
>>  
>> Richard Kingston, PhD.
>> School of Biological Sciences
>> The University of Auckland
>> New Zealand
>>  
>> website: http://persephone.sbs.auckland.ac.nz/richard/lab/
>>  
>>  
>> 
>>  
>> _______________________________________________
>> RASMB mailing list
>> RASMB at list.rasmb.org
>> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
> 
> 
> This message and any attachment are intended solely for the addressee and may contain confidential information. If you have received this message in error, please send it back to me, and immediately delete it.   Please do not use, copy or disclose the information contained in this message or in any attachment.  Any views or opinions expressed by the author of this email do not necessarily reflect the views of the University of Nottingham.
> 
> This message has been checked for viruses but the contents of an attachment may still contain software viruses which could damage your computer system, you are advised to perform your own checks. Email communications with the University of Nottingham may be monitored as permitted by UK legislation.
> 
> 
> _______________________________________________
> RASMB mailing list
> RASMB at list.rasmb.org
> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org