[RASMB] Absorbance Scan Question
Matthew Hockin
mhockin at gmail.com
Tue Jan 31 09:26:06 PST 2012
I think your ratio is wrong for this to be effecting your data... so i make this comment because I am not absolutely sure about which way you did the ratio and this may possibly play into it- the Xenon spectrum has a pretty stark trough right at 240.. whereas 236 would be much more intense (as much as 2-3 fold). If my spectrum is correct. Its possible that with different enough input intensity you could be seeing non-linear behavior in one data set (remote...). It would be just my luck to do this exact experiment and run into something like this so I bring it up.
Matthew Hockin
On Tue, Jan 31, 2012 at 10:12 AM, Walter Stafford <stafford at bbri.org> wrote:
This could be a monochromater alignment problem. If the m.c. is out of alignment as the scan proceeds the wavelength will vary from one end of the scan to the other ad that could account for the varying ratio.
==============================================
Walter F. Stafford, Ph.D.
Senior Scientist and
Director Analytical Ultracentrifugation Research Laboratory
Boston Biomedical Research Institute
64 Grove Street
Watertown, MA 02472
617-658-7808
skype: w.stafford3
----------------------
"Be kind, for everyone you meet is fighting a hard battle." -Plato
------------------------------------
On Jan 31, 2012, at 11:32, Peter Edward Prevelige Jr wrote:
> Hi All –
>
> I’m doing an equilibrium run on a ~3 KDa synthetic peptide lacking aromatics in 20% TFE. I thought I might use multiple wavelengths to extend the concentration range that I could follow.
>
> I thought I’d be clever and obtain the ratio of extinction coefficients by dividing the absorbance at one wavelength (236 nm) by the other (240 nm) across the cell (rather than using the spectrum).
>
> I calculated the ratio using only the points that were at the same radial distance and plotted them vs radial distance. It is pretty apparent that the ratio is changing across the cell.
>
> My question is whether this is likely to be an optical artifact or whether it most likely represent heterogeneity in the sample. It is probably worth noting that the ratio from the spectrum is 1.43, intermediate between the extremes seen across the cell which to me suggests heterogeneity.
>
> Thanks
>
> Peter
>
> Peter E. Prevelige Jr.
> Professor
> Dept. of Microbiology, BBRB 416/6
> Univ. of Alabama @ Birmingham
> 845 19th St. South
> Birmingham AL. 35294-2170
> Phone 205 975-5327
> FAX 205 975-5479
> prevelig at uab.edu
> http://www.microbio.uab.edu/faculty/prevelige/prevelige-p.htm
>
> <Scan.pdf>_______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
_______________________________________________
RASMB mailing list
RASMB at rasmb.bbri.org
http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20120131/bb7d57ae/attachment-0001.html>
More information about the RASMB
mailing list