[RASMB] error estimates and F-statistics in Sedphat
John Philo
jphilo at mailway.com
Tue Apr 26 13:42:42 PDT 2011
Mark,
Given that the fit is so bad I'm not sure it makes any sense to try to get a
confidence interval. The model is clearly not even coming close to
representing the data, so what does this association constant mean?
I agree fully with Jack that you cannot expect the peaks from c(s) analysis
to give you the monomer and dimer sedimentation coefficients, and that you
really want to do a dilution series over a big concentration range. For
absorbance data that usually means either using multiple wavelengths (in
different runs) or using the 3 mm centerpieces to help cover a big range
while keeping the OD <=1 for the high end. Then you can either fit the
weight-average sedimentation coefficient versus concentration to a model, or
do global whole boundary analysis with SEDPHAT or SEDANAL (with the
sedimentation coefficients as fitting variables).
For this or any other experiment I also recommend that you fill you cells
more completely to get a better separation. Try to get the meniscus at
5.9-5.95 cm.
John
_____
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Mark Agacan
Sent: Tuesday, April 26, 2011 8:36 AM
To: RASMB at rasmb.bbri.org
Subject: Re: [RASMB] error estimates and F-statistics in Sedphat
Thanks Patrick and Chad for the information about error estimates from
F-statistics.
My problem now is that the apparent errors for the binding constants derived
from the analysis of F-statistic appear to be very large. When I make a
change to e.g. log10Ka12, by any value from 0.1 - 3.0, there is no change
(or only a very small change) in chi squared, so I can never reach the
critical chi squared.
Each dataset looks good and gives low rmsd values when processing with the
c(s) model in sedfit. I'm using the monomer-dimer self-association model in
sedphat, with the values for S1 and S2 taken from c(s) analysis. I have
tried to fit to other models with only worse results.
The fit looks very poor (see screenshots attached). I don't think I need to
repeat the experiment because the data looks good, at least to my eye, but
the fit to the monomer-dimer model seems very bad.
Any further suggestions or advice would be very welcome.
Many Thanks,
Mark
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr Mark Agacan, Scientific Officer for the Division of Biological Chemistry
and Drug Discovery,
Wellcome Trust Biocentre, College of Life Sciences, University of Dundee,
Dundee, DD1 5EH
Tel: +44 1382 386095 Fax: +44 1382 345764 Mobile: 07525 451 117
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
************************************************************
Please consider the environment. Do you really need to print this email?
>>> Chad Brautigam <Chad.Brautigam at UTSouthwestern.edu> 4/23/2011 15:02 >>>
Hi, Mark,
The F-statistics in SEDPHAT are meant to allow you to construct a
"do-it-yourself" error interval. Check out the statistics help page. Under
the heading "Error Analysis" there is a procedure for how to use the output
"critical chi-square" from the Statistics Menu to find an error interval.
Cheers,
Chad
_____
From: rasmb-bounces at rasmb.bbri.org [rasmb-bounces at rasmb.bbri.org] on behalf
of Mark Agacan [M.Agacan at dundee.ac.uk]
Sent: Friday, April 22, 2011 6:36 AM
To: rasmb at rasmb-email.bbri.org
Subject: [RASMB] error estimates and F-statistics in Sedphat
Hello,
I'm calculating errors in the binding constants in sedphat as instructed in
the statistics help page, but I don't appear to get an output for the
F-statistics. Are the F-stats output only as (changes in) chi-squared
values? e.g. the error associated with the binding constants is clearly
given in the covariance matrix.
Best Wishes,
Mark
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr Mark Agacan, Scientific Officer for the Division of Biological Chemistry
and Drug Discovery,
Wellcome Trust Biocentre, College of Life Sciences, University of Dundee,
Dundee, DD1 5EH
Tel: +44 1382 386095 Fax: +44 1382 345764 Mobile: 07525 451 117
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
************************************************************
Please consider the environment. Do you really need to print this email?
The University of Dundee is a registered Scottish charity, No: SC015096
_____
UT Southwestern Medical Center
The future of medicine, today.
The University of Dundee is a registered Scottish charity, No: SC015096
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20110426/fedee0ab/attachment.htm>
More information about the RASMB
mailing list