Dear RASMBers,

Thanks to Virgil, Andrew, Stephen, Karen and Sue for replying to my message last week concerning the tripping problem with our XL-I.  

It has just been fixed by our (very helpful) Beckman engineer, and the problem was different from all the possibilities suggested in the replies.  

Here's how he fixed it:

He reckoned that the problem was likely to reside with the battery backed RAM in the system control board - this can apparently become corrupted occasionally, causing the circuit breaker to trip.

So he swapped the battery backed RAMs on the CRT and system control boards and the problem was resolved (as far as we can tell).

If you should ever have this problem, this fix is not the kind of thing to try yourself; I'm relaying this to RASMB so that if it does happen to you, you know that a (brief) engineer visit is needed and you know what your engineer might try!

All the best,
Olwyn


On 21 Mar 2011, at 13:06, Olwyn Byron wrote:

> Dear RASMBers,
> 
> I'm writing to ask whether any on you can tell me what might be wrong with our XL-I.
> 
> When one of our students was setting up an SV run yesterday he ran into the following problem:
> 
> The XL-I was set to accelerate to 49 krpm in the XLA 4.5 software.
> When the "start single scan" button was pressed (for initial set-up) the XL-I accelerated a bit (to about 20 krpm) and the student noticed that on the control panel the speed set was 48 krpm, not 49 krpm.
> 
> Then an error message appeared on the XL-I control screen - which was flashing - "mismatch in configuration between centrifuge and computer - settings adjusted accordingly - press any key to remove message".
> 
> Pressing any key gave an alert sound and the message remained.
> 
> So the student hit the stop button on the XLA 4.5 software control screen; the XL-I stopped.  He turned the vac off, opened the lid, removed the rotor and optics.
> 
> Then he switched the XL-I off.
> 
> Then he couldn't turn it back on.  
> 
> In fact what happens is when we flick the "on" switch, power comes on for 3 seconds and then trips off.
> 
> Has anyone had this type of experience before?  If so, could you advise what we could try to get the XL-I back on?
> 
> Many thanks and best wishes to all-
> Olwyn
> 
> ________________________________________________________________
> Olwyn Byron
> 
> ****please note new email address: olwyn.byron at glasgow.ac.uk****
> 
> B2-11, Glasgow Biomedical Research Centre, University of Glasgow
> 120 University Place, Glasgow G12 8QQ, Scotland, UK
> 
> +44 (0)141 330 3752
> olwyn.byron at glasgow.ac.uk
> http://www.gla.ac.uk:443/ibls/staff/staff.php?who=Pn||ed
> ________________________________________________________________
> The University of Glasgow, charity number SC004401
> 
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb




More information about the RASMB mailing list