Just a correction: from fluorescence anisotropy you don't get the radius of
gyration, but possibly the harmonic mean of the rotational correlation time.
As for using TEM with negative staining, one has to be absolutely sure that
adhesion on the grid doesn't influence the monomer-dimer equilibrium...
Mattia
At 08:52 AM 1/27/2011 -0500, Titus M. Franzmann wrote:
>Ewa,
>Typically one tries to span a protein concentration range of at least 5times
>lower and 5times greater the kDa. Optimally 10times.
>So you would go at least from 2.5 to about 125 nM. Given the already low
>signal intensity of 0.1 at 220 at 25 nM you may not be able to observe the
>fully dissociated particle (monomers) and thus may not be able to acquired a
>dataset with a fully established monomer baseline. That would give an
>uncertainty in the Kd when fitting it to a mono-dimer association model
>(assuming no equilibrium intermediate populates).
>Anyway, I have carried out a couple of experiments at low signals, e.g. 0.05
>Abs. and determining the S-value is fairly robust, but fitting these low
>signal datasets to find proper Mw values is fairly difficult. Of course one
>can gain confidence by running triplicates and consolidate the datasets in a
>global fit. From this you can at least estimate what kind of particles you
>have. If the system is a fast interacting system, you should be able to use
>the Sw20avg values and plot them as the function of the protein
>concentration. It will be difficult to do this, when the system is slow
>interacting and the signal gets separated and distributed among the monomer
>and dimer fractions.
>Anyway, have you considered fluorescence anisotropy? According to your
>information, the protein contains several trp residues that you can use to
>monitor the assembly process. From the fluorescence anisotropy you can also
>gain information about the radius of gyration and make some estimate about
>the shape and size. Alternatively, we have recently used TEM to visualize
>monomers and dimers of a 39 kDa monomer. Surprisingly, we got some
>information using negative staining techniques and particle population
>analysis. You would at least get some idea about what you get.
>Titus
>
>Dr. Titus M. Franzmann
>S. G. Walter Lab
>Mol., Cell. and Develop. Biol. Dept.
>University of Michigan
>4140C Nat. Sci. Bldg
>830 N. University Ave.
>Ann Arbor, MI 48109-1048
>
>
>-----Original Message-----
>From: rasmb-bounces at server1.bbri.org [mailto:rasmb-bounces at server1.bbri.org]
>On Behalf Of Leech, AP
>Sent: Thursday, January 27, 2011 8:17 AM
>To: rasmb at server1.bbri.org
>Subject: Re: [RASMB] AUC capabilities
>
>Hi All,
>
>Eva, if you put all 50 ug into 120 uL for an equilibrium experiment,
>this would be ~0.4 mg/ml, or 7 uM, giving an approx absorbance of
>0.4 at 280 nm. If your Kd is right, this would practically all be
>dimer. To get down to sub-Kd concentrations, as Borries says, the
>absorbances would be impractically low.
>
>If you put it all into 400 uL for a velocity experiment, the initial
>concentration would be 0.125 mg/ml, 2.1 uM, absorbance ~0.13 at 280nm.
>This is well above the Kd also, I will leave someone with more
>experience to say if dissociation would affect the boundary detectably,
>I suspect not.
>
>It might be worth considering SPR. If you can attach one monomer to
>the chip surface without compromising dimerisation (practically I think
>you would attach the dimer and look for dissociation; there might be
>a problem with both halves of the dimer attaching), then wait for
>dissociation and flow various concentrations over the surface. The
>amounts are more in the ball park for what you have (say 10 ug for
>immobilisation and a few ug for each trial of free protein) though
>it's still not a lot! and you need to allow for control experiments
>for non-specific binding etc.
>
>Best regards,
>
>Andrew
>
>On 27/01/2011 04:29, Ewa Folta-Stogniew wrote:
> > At 10:39 PM 1/26/2011, you wrote:
> >> >
> >> > Hello,
> >> >
> >> > is it possible to determine dimerization constant in the order of 25
> >> > nM using AUC? What will be the minimal amount of protein needed?
> >> >
> >> > Ewa
> >> >
> >>
> >> Hi Ewa,
> >>
> >> the answer to your question is: it depends on several factors...
> >>
> >> 1. what optical system do you plan to use - there are choices:
> >> absorbance, Rayleigh interference, fluorescence emission
> >
> > absorbance or interference
> >
> >
> >> 2. If you use absorbance, there is a question about the extinction
> >> coefficient and the wavelength. Depending on extinction, 25 nM may
> >> be well within reach at several detection wavelengths.
> >
> > A280(0.1%)=0.85 (mg-1mL cm-1)
> >
> >> 3. Related to (2) is the question of size - the larger the protein,
> >> presumably the larger the extinction coefficient (more trp, tyr residues
> >> at 280 nm, more peptide backbone absorbance at lambda <= 230 nm). Ditto
> >> for Rayleigh interference - the larger the protein, the more refraction
> >> you will see and 25 nM could be well within range.
> >
> >
> > 59 kDa monomer
> >
> >
> >> 4. If you consider using the fluorescence detector, then 25 nM label
> >> of the right excitation/emission should be well within the range of
> >> detection.
> >
> > no fluorescence detection available
> >
> >
> >> 5. in order to get accurate Kds you would want to measure ideally
> >> at concentrations above and below the Kd to get good signal
> >> of the monomer and the dimer, so multiple concentrations measured
> >> either in velocity or equilibrium mode and globally fitted should
> >> give you the desired information. Related to that is that the
> >> Kd needs to be in the range where you are measuring. If your Kd is
> >> 100 micromolar and you are measuring at 25 nM and below, your answer
> >> is going to be quite unreliable. At the same token, measuring at
> >> 100 uM when your Kd is at 20 nM will not give you the desired result.
> >
> > Assuming Kd of 25 nM and the extinction coefficient as above- can one
> > determine Kd for dimerization from 50 micrograms of protein total (25
> > micrograms preferred)?
> >
> > If this is theoretically feasible, does anybody know of any published
> > AUC results for similarly low Kd?
> >
> > thanks,
> >
> > Ewa
> >
> >
> >> Your question about how much protein is needed at a minimum depends
> >> also on the questions I posed above. If you provide additional
> >> information we may be able to tell you more specifically for your sample.
> >>
> >> Regards, -borries
> >
> > _______________________________________________
> > RASMB mailing list
> > RASMB at rasmb.bbri.org
> > http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
>
>--
>Dr Andrew Leech * Laboratory Head
>Technology Facility * Molecular Interactions Laboratory
>Department of Biology (Area 15) * Tel : +44 (0)1904 328723
>University of York * Fax : +44 (0)1904 328804
>PO Box 373, York YO10 5YW * Email : apl3 at york.ac.uk
>EMAIL DISCLAIMER: http://www.york.ac.uk/docs/disclaimer/email.htm
>_______________________________________________
>RASMB mailing list
>RASMB at rasmb.bbri.org
>http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
>
>
>_______________________________________________
>RASMB mailing list
>RASMB at rasmb.bbri.org
>http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
ATTENZIONE.
Il presente messaggio ed i suoi allegati devono intendersi ad uso esclusivo
dei suoi destinatari e sono confidenziali. Persone diverse dal
destinatario, anche ai sensi del D.Lgs. n. 196/2003 , non debbono prendere
visione dei contenuti, copiare od inoltrare l'e-mail a terzi. Se ricevete
questo messaggio per errore, Vi preghiamo di cancellarlo, di distruggerne
ogni copia e di informarci immediatamente. La mail non è un protocollo
sicuro, pertanto IST declina ogni responsabilità in caso di intercettazione
o modifiche del presente messaggio.
WARNING.
This message and any attachments is intended solely for the use of the intended
addressees and is confidential. Any unauthorized review, use, disclosure or
distribution
is prohibited. If you receive this message in error, please delete it,
destroy all copies
and immediately notify us. Mail is not a secure protocol, therefore IST
will not be liable
for interception or amendment of this message.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20110127/801e511b/attachment.htm>
More information about the RASMB
mailing list