[RASMB] Fwd: Re: AUC capabilities
Ewa Folta-Stogniew
ef55 at email.med.yale.edu
Thu Jan 27 06:29:26 PST 2011
>Date: Thu, 27 Jan 2011 09:19:28 -0500
>To: Allen Minton <minton at helix.nih.gov>
>From: Ewa Folta-Stogniew <ef55 at email.med.yale.edu>
>Subject: Re: [RASMB] AUC capabilities
>
>Allen,
>
>thanks for reminding me... I only thought about this approach for
>high concentrations..
>
>Is everything I may need for this approach (assuming minimal sample
>volume) commercially available?
>
>Ewa
>
>At 08:46 AM 1/27/2011, Allen Minton wrote:
>>At 09:41 PM 1/26/2011, you wrote:
>>>Hello,
>>>
>>>is it possible to determine dimerization constant in the order of
>>>25 nM using AUC? What will be the minimal amount of protein needed?
>>>
>>>Ewa
>>
>>The answer is yes, but you will have to employ the non-traditional
>>approach we use in our lab, which involves centrifugation to sed.
>>eq., followed by fractionation of the cell contents (see attached
>>reprints). Then you can use any technique available to measure the
>>relative concentration of the protein in each fraction and thereby
>>reconstruct the equilibrium gradient. There are many analytical
>>methods for measuring protein concentration at much lower
>>concentrations accessible by either the absorbance or RI scanner in
>>your analytical ultracentrifuge, such as the widely used BCA
>>colorometric assay. If you are willing to radiolabel your protein
>>or to employ an immunoadsorbent assay you can go one or two orders
>>of magnitude even lower in concentration.
>>
>>I realize that this would require you to learn a whole new way of
>>doing SE measurements, but it would provide you with a lot of new
>>capabilities. For example, we are now using this method to study
>>associations of dilute proteins in solutions containing high
>>concentrations of other macromolecules. Can't do that with your Beckman!
>>