[RASMB] determination of phage particle concentration
John Correia
jcorreia at biochem.umsmed.edu
Mon Mar 8 07:29:51 PST 2010
Berkowitz and Day Biochemistry 1980 19, 2696-2702. Turbidity
Measurements in an Analytical Ultracentrifuge: Determinations of Mass
per Length for Filamentous Viruses fd, Xf and Pf3. Combines scattering
and interference measurements in a Model E. Can it be done in an XLI is
the question.
>>> On 3/8/2010 at 8:08 AM, in message
<20100308141543.BE7051742AF at rasmb.bbri.org>, Glen Ramsay
<glen at avivbiomedical.com> wrote:
Greetings RASMB:
The scattering increases as the fourth power of the inverse wavelength.
Years ago Aviv's spectrophotometer had a function that would fit the
baselines before and after an absorbance peak to subtract the scattering
portion.
However, I cringe at the idea of making analytical measurements of
particles via absorbance. Absorbance of a photo can happen only once.
Scattering can happen more than once, sending a photon back into the
detector's view. The detectors of different instruments view different
collection angles of the sample, so the amount of scattered light is
going to vary depending on the instrument design. Reflections of
scattered light off the surroundings can send photons back onto the
detector. In one custom instrument Aviv mounted the detector on a rail,
and apertures were mounted on both the sample's exit and detector's
entrance, to control the light collection angle. And of course the size
of the particle affects the amount of light scattering.
Measure scattering with absorbance with both eyes and mind open,
because the instrument is not being used as intended.
Glen
At 09:44 AM 3/5/2010, Steve Harding wrote:
Content-class: urn:content-classes:message
Content-Type: multipart/alternative;
boundary="----_=_NextPart_001_01CABC72.57041806"
Dear Sabine
Simple turbidity measurements (making sure you are away from absorption
maxima) using a good quality spectrophotometer may suffice for the sort
of information you are after - particle concentration, although
stricvtly speaking it is not a hydrodynamic method.. This is is the
usual method for measuring concentrations (particles per ml) of very
large assemblies such as spores and I think it works for smaller
assemblies that are not too dilute..
Victor Bloomfield's group did quite a bit on the turbidity of phages in
the late 70's and gave the relevant formulae (you may need the approx
mol. wt value) - if you look up Bahls and Bloomfield (and not Victor's
QLS papers) that should give you a lead!
All best
Steve Harding
http://www.nottingham.ac.uk/ncmh
> Dear all,
>
> we are trying to obtain the concentration of bacteriophages in a
> solution, concentration meaning the number of particles per ml. The
> solution is supposed to be monodisperse and the MW is (roughly)
known
> but could be determined exactly. Is the any idea of how to do this
> with a hydrodynamic method (staining is difficult to compare between
> phage mutants and counting infected cells is tedious). I`ve got the
> "feeling" that one could use the number-averaged MW, but maybe this
is
> totally wrong.
>
> Thank you for any hints, literature is also welcome.
>
> Cheers,
>
> Sabine
>
>
>
>
> Sabine Kaltofen
> PhD student
>
> Universität Potsdam
> Department of Physical Biochemistry
> Institute of Biochemistry and Biology
> Karl-Liebknecht-Str. 24-25, Haus 25, Raum B/0.05
> D-14476 Potsdam-Golm
> Telefon: +49-(331)-977-5245
> Email: kaltofen at uni-potsdam.de
>
>
This message has been checked for viruses but the contents of an
attachment may still contain software viruses which could damage your
computer system: you are advised to perform your own checks. Email
communications with the University of Nottingham may be monitored as
permitted by UK legislation.
_______________________________________________
RASMB mailing list
RASMB at rasmb.bbri.org
http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
Glen Ramsay, Ph.D.
Chief Scientist
Aviv Biomedical, Inc.
750 Vassar Avenue, Suite #2
Lakewood, NJ 08701
(732) 370-1300, Ext. 25
(732) 370-1303, FAX
glen at avivbiomedical.com
www.avivbiomedical.com
( http://www.avivbiomedical.com/ ) Skype: GlenAtAviv
The information transmitted is intended only for the person or entity
to which it is addressed and may contain confidential and/or privileged
material, the disclosure of which is governed by applicable law. Any
review, retransmission, dissemination or other use of, or taking of any
action in reliance upon, this information by persons or entities other
than the intended recipient is prohibited. If you received this in error
please contact the sender and destroy the materials contained in this
message.
Individuals who have received this information in error or are not authorized to receive it must promptly return or dispose of the information and notify the sender. Those individuals are hereby notified that they are strictly prohibited from reviewing, forwarding, printing, copying, distributing or using this information in any way.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20100308/b5964282/attachment-0001.html>
More information about the RASMB
mailing list