Sorry, I forgot to say earlier with respect to Liz's suggestion that indeed
unless you are careful switching to a new computer will  lead to a loss of
instrument calibration and incorrect sedimentation coefficients and masses.
The radial and wavelength calibration data are stored in two files in the
root directory of the instrument computer (calparm.cal and waveparm.cal). If
these files are not manually transferred to the new computer you will lose
the calibration.
 
John

  _____  

From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of John Philo
Sent: Thursday, November 05, 2009 8:09 AM
To: RASMB
Subject: Re: [RASMB] N/A


Kanchan, I suggest you search the archives, there have been many discussions
of this topic. Briefly, with an XL-A you can pretty much use any computer
and version of Windows, but with an XL-I it depends on which version of the
camera you have. If your old XL-I computer was running Windows 98 then you
MUST stick with Windows 98 (or pay Beckman a huge sum to update the camera
and software). Even if you do have the newer camera and were running Windows
2000 or XP, the camera interface (Raptor) board is very fussy about the
computers and will likely not work in this new computer. Your new Dell also
probably does not even have the serial port needed to connect to the XL-I. 
 
In short, you are probably better off sticking to your old computer. There
should not be any memory issue if you are strictly using it for data
acquisition. If you are out of hard drive space then simply use the CD
burner to transfer the data elsewhere. It is a bad idea to use the XL-I
computer for data analysis anyway since you are likely to crash the ongoing
experiments.
 
John

  _____  

From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Bartlett, Lizz
Sent: Thursday, November 05, 2009 5:26 AM
To: 'Kanchan Gupta Phd Batch 2006 Cell Bio R.P.Roy'; RASMB at server1.bbri.org
Subject: Re: [RASMB] N/A



Hello,

I would suggest that you do a few experiments with BSA or some other protein
where you know very well your s values. I know of a lab that ran into
problems when they switched computers. There were COM errors that delayed
the time stamp for each scan. 

 

Good Luck,

 

 

Lizz Bartlett

Scientist I

ImmunoGen, Inc.

830 Winter St.

Waltham, MA 02451

Voice: 781.895.0600

www.immunogen.com

  _____  

From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Kanchan Gupta Phd Batch 2006 Cell Bio R.P.Roy
Sent: Thursday, November 05, 2009 1:35 AM
To: RASMB at server1.bbri.org
Subject: [RASMB] N/A

 

hi

we were facing problems because of low memory of our old Pentium 3 computer.
Therefore, we ordered a new Dell (TM) Inspiron (TM) 545 desktop with windows
xp 32 bit, 2.93GHz processor and 4 GB RAM. i wanted to know if the new
computer can be used with the Beckman Optima XL-I analytical
ultracentrifuge, class T instrument. It was installed in 2002-03 and the
serial no. is W103A02. The operating software version available is Beckman
XL-x (BECKMA~1.LNK).

regards


Kanchan Gupta 
Research Scholar 
Cell Biology Lab-II 
National Institute of Immunology 
Aruna asaf ali marg 
New Delhi-110067

INDIA 

---------- Original Message ----------- 
From: John_Doran at vrtx.com 
To: RASMB at server1.bbri.org 
Sent: Wed, 4 Nov 2009 19:26:27 -0500 
Subject: [RASMB] Thanks to everyone who took the time to reply to my query 

> All the responses were very useful, thank you again 
> 
> John D Doran, Ph.D. 
> Protein Biochemistry 
> Vertex Pharmaceuticals Inc. 
> 130 Waverly Street 
> Cambridge, MA 02139 
> 617-444-6142 
------- End of Original Message ------- 

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20091105/19873b93/attachment.htm>


More information about the RASMB mailing list