[RASMB] Small peptides and AUC
Jo Butler
pjgb at mrc-lmb.cam.ac.uk
Tue Apr 29 06:18:47 PDT 2008
Dear All,
On v-bar Dave raises a very valid point. There is however one major
problem with trying to match densities with the solvent and this is that
lipids and detergents usually do not have the same v-bar themselves, so
it is not possible to match both in the same experiment. This has often
been overlooked in density-matching experiments.
If I may push a personal opinion, measuring the density increment
directly has great advantages and is the one really reliable method. We
discussed the whole problem in Butler, P. J. G. & Tate, C. G. (2005)
Correcting for the Buoyancy of Macromolecules; Density Increments and
Apparent Partial Specific Volumes with particular reference to the study
of membrane proteins in Analytical Ultracentrifugation: Techniques and
Methods (Scott, D., Harding, S. & Rowe, A., eds) pp. 133-151, Royal
Society of Chemistry, London.
Yours,
Jo
Scott David wrote:
>
> Hello All,
> Following on from discussions of 90 % DMSO,etc. There are more
> applications (at least through the NCMH) to look at peptides in
> membranes and detergents, and talking to people in the AUC field, this
> is a somewhat growing field as people become less fearful of
> membrane-bound stuff. At this juncture I think it would be good if we
> could start a discussion strand of what to watch out for when
> analysing these samples. Original references very welcome.
>
> Here's several very obvious things to get us started.
>
> 1) vbar. The bouyancy of the peptide changes with binding of lipids
> and detergent. This has to be either (a) matched out by having a
> solvent that has the same density as the lipid/detergent or (b)
> calculated from the density increment. I also had a referee once ask
> me years ago whether I had corrected for charge on a small peptide.
> The answer was "no" as the peptide was actually a reasonably sized
> protein, but the original reference for this would be good.
>
> 2) Speed. Peptides are small, you will need probably the highest speed
> possible to analyse them, and for the XL-A/I that's 60K and the
> associated problem of buying the correct centrepieces (Aluminium or
> SpinAnalyticals SedVel60K).
>
> 3) Co-sedimentation of the solvent. Peptides are small, therefore have
> low S values. I seem to recall that Na^+ sediments at around 0.3 S,
> therefore you will need to think about how the sedimentation of your
> solvent components affect the sedimentation of your sample. Not
> trivial if these affect the solubility and stability of your peptide.
>
> 4) Small S values. Are our algorithms good enough?
>
>
> Enjoy!
>
> Dave.
>
>
>
> Dr. David J. Scott
> Lecturer in Physical Biochemistry
> National Centre for Macromolecular Hydrodynamics
> School of Biosciences
> University of Nottingham
> Sutton Bonington Campus
> Sutton Bonington
> Leicestershire
> LE12 5RD
> United Kingdom
> dj.scott at nottingham.ac.uk
> _www.nottingham.ac.uk/ncmh_ <file://www.nottingham.ac.uk/ncmh> _
> +44 (0) 115 951 6221 Phone
> +44 (0) 115 951 6142 Fax _
>
> _
> _
>
> _This message has been checked for viruses but the contents of an
> attachment may still contain software viruses, which could damage your
> computer system: you are advised to perform your own checks. Email
> communications with the University of Nottingham may be monitored as
> permitted by UK legislation. _
>
> _
> _
> ------------------------------------------------------------------------
> _
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
> _
--
Dr P.J.G. Butler,
MRC Laboratory of Molecular Biology,
Hills Road, Cambridge, CB2 0QH, UK.
Tel. +44 (0)1223 402296
More information about the RASMB
mailing list