[RASMB] upper concentration limit AUC
Tom Laue
tml at cisunix.unh.edu
Mon Sep 8 05:19:33 PDT 2008
Hi-
Food for thought and a serious question.
The "natural" concentration units for k(s) are volume fraction (since it
is a hydrodynamic term).
The "natural" concentration units for k(D), or k(P) in Arthur's
discussion, are mole fraction (since it is a thermodynamic term).
Because of the different "natural" concentration units, is there a
fundamental difficulty in using sedimentation velocity to analyze high
concentration solutions?
Best wishes,
Tom
Arthur Rowe wrote:
>
> Hi Christine (and everyone)
>
> I think this discussion is already in the RASMB Archives.
>
> To summarise very briefly, k(D) is indeed lower than k(s), but not
> hugely so. For simple, spherical particles, defining all
> coefficients in volume fraction terms and at limiting infinite
> dilution, we can write
>
> k(D) = k(P) - k(s) where k(P), the concentration dependence of the
> chemical potential which is the force driving the flux in
> translational diffusion, and is given for s single component by
> (2BM + higher terms) where B is the 2nd virial coefficient. For
> spheres, 2BM = 8 ml/g; k(s) = 4* ml/g: so k(D) = 4 ml/g. I am
> afraid that neglecting k(D) is not a good approximation.
>
> Kind regards - will be seeing you (Christine) and lots of others
> in Newcastle this week.
>
> Arthur
>
> *see "The sedimentation rate of disordered suspensions" Brady,
> John F.; Durlofsky, Louis J.
> Physics of Fluids 1988 31 717-727 for the last word from the fluid
> mechanics people. You can also look up my Chapter in the 1992 Book
> (the 'black book') ed s Harding, Rowe & Horton) for details of my
> own derivation. Which happens to be in total numerical agreement
> with the Brady/Durlovsky treatment, right p to 64% volume fraction
> . . . . .
> *Note: *the value given above for k(s) = 4 ml/g is for *DYNAMIC
> k(s)*. The usual measured value which gets reported is the
> *KINEMATIC* k(s), which for spheres has the value 5.0 ml/g. You
> can get either value out of theory, depending how you play it.
> Brady & Durlovsky report the kinematic value. All of which was
> first noted by Burgers in 1939/40, and has been much ignored ever
> since! As no density 'correction' is called for in the estimation
> of k(D) or of k(P), it seems reasonable to use the dynamic k(s) in
> the equation above.
>
>
> Dear Joris, dear all,
>
> This reference may be perhaps useful.
>
>
>
> Solovyova A., Schuck P., Costenaro L., Ebel C,
>
> Non-ideality by sedimentation velocity of halophilic malate
> dehydrogenase in complex solvents,
>
> (2001) /Biophys. J,/ 81 1868-1880.
>
>
>
> In this work, we analyzed sedimentation velocity profiles
> considering hydrodynamic and thermodynamic non ideality. (i.e.
> concentration dependency of s and D) in the case of an
> homogeneous solution of our protein of interest. The modified
> Lamm equation was implemented in a model of analysis in Sedfit
> (note that, unless the programme was recently modified, the kS and
> kD are expressed in signal unit in sedfit)
>
> From my experience, the sedimentation velocity profiles are
> essentially modified by the concentration dependency of the
> sedimentation coefficient. Thus the concentration dependency of D
> can be neglected in a first approximation. Also from a theoretical
> point of view, kD is much lower that ks.
>
> All the best
>
> Christine
>
>
> Christine EBEL
>
> Institut de Biologie Structurale CEA-CNRS-UJF
>
> 41 rue Jules Horowitz, F-38027 Grenoble France
>
> Tel (33) (0) 4 38 78 95 70; Fax (33) (0) 4 38 78 54 94
>
> christine.ebel at ibs.fr
>
> http://www.ibs.fr/content/ibs_eng/presentation/lab/lbm/ebel.htm
> -----Message d'origine-----
> *De :* rasmb-bounces at rasmb.bbri.org
> [mailto:rasmb-bounces at rasmb.bbri.org] *De la part de* Beld, Joris
> *Envoyé :* lundi 1 septembre 2008 15:24
> *À :* RASMB at rasmb.bbri.org
> *Objet :* [RASMB] upper concentration limit AUC
>
>
>
> Dear all,
>
>
>
> A colleague asked me whether analytical ultracentrifugation has an
> upper limit with regard to the concentration of the protein. They
> want to measure the protein at the same concentration as the NMR
> experiments (> 1 mM). I am not entirely sure but I thought this
> should be no problem. One could easily measure off-peak at another
> wavelength than 230nm, e.g. 235nm or 280nm, right?! Or does one
> run into non-ideality phenomena when doing sedimentation
> equilibrium at these high protein concentrations?!
>
> Thanks a lot in advance for any feedback.
>
> Best wishes,
>
>
>
> Joris Beld
>
>
>
> Hilvert Group
>
> ETH Zürich
>
> Switzerland
>
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
>
>
>
> This message has been checked for viruses but the contents of an
> attachment may still contain software viruses, which could damage your
> computer system: you are advised to perform your own checks. Email
> communications with the University of Nottingham may be monitored as
> permitted by UK legislation.
>
> ------------------------------------------------------------------------
>
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
>
--
Department of Biochemistry and Molecular Biology
University of New Hampshire
Durham, NH 03824-3544
Phone: 603-862-2459
FAX: 603-862-0031
E-mail: Tom.Laue at unh.edu
www.bitc.unh.edu
www.camis.unh.edu
More information about the RASMB
mailing list