AW: [RASMB] Interpreting SV and SE data
Eric Salgado
esalgado at popmail.ucsd.edu
Thu May 29 12:04:04 PDT 2008
Titus,
My ranges are from 5 uM to 1 mM ( viscosity at the higher end
doesn't seem to be a problem).
Titus M. Franzmann wrote:
> Hi Eric,
> may I ask what the concentration range is you cover in your experiments?
> Best regards
> Titus
>
> -----Ursprüngliche Nachricht-----
> Von: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] Im
> Auftrag von Eric Salgado
> Gesendet: Donnerstag, 29. Mai 2008 19:55
> An: rasmb at server1.bbri.org
> Betreff: [RASMB] Interpreting SV and SE data
>
> Hello all,
> I have a system for which I've done SV experiments at multiple
> concentrations that, on increasing the concentration, reveals a peak
> that starts at an s value of about 2, shifts to a plateau between 2.4
> and 2.6, then steadily increases at 2.6. I should point out that all of
> these peaks are very broad.
> The c(MW) analysis describes a monomer at low concentrations; dimer
> at intermediate concentrations; something between dimer and trimer at
> high concentrations; and finally, as an aggregate starts to form at
> very high concentrations, a trimer peak is resolved. This trimer peak
> correlates to our crystal structure, but it appears as if, because of
> the apparently low association constant and fast kinetics involved with
> the trimer formation, this fact is masked in the SV data.
> In order to hopefully rid of this ambiguity, I performed SE
> experiments in the concentration regime that did not form an aggregate
> (and therefore apparently only a small amount of trimer), but find that,
> as far as a species analysis goes, both a model with monomer-dimer or
> monomer-dimer-trimer present are statistically equal. That being said,
> the statistics are lower for the later model, while the fit and
> residuals for both are equally good.
> With all of this in hand, how can I determine the best SE model to
> use, and can I say with certainty that the trimer exists in solution
> from these data?
>
> I know this is a lot, but I would appreciate any help I can get with
> this problem.
>
> Thank you in advance,
>
> Eric Salgado
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
>
>
More information about the RASMB
mailing list