[RASMB] k_off rate
SE Halford, Biochemistry
S.Halford at bristol.ac.uk
Mon Apr 7 07:09:41 PDT 2008
Sorry, but I disagree strongly with Allen Minton's assertion that
protein-protein association rates have "very similar bimolecular
association rate constants (say within a factor of 3 or so). I seem to
remember something on the order of 10^5 mol^-1 sec^-1".
The protein-protein association rates reported in the literature vary from
10^10 M^-1.s^-1 down to around 10^4 M^-1.s^-1. The high numbers are almost
all measured in low salt buffers, conditions where electrostatic effects
overwhelm the "Smoluchowski limit" of around 10^8 M^-1.s^-1, and when the
same protein-protein interaction is measured in buffer containing ~250 mM
NaCl, the rate falls to ~10^8 M^-1.s^-1, somewhere close to the
"Smoluchowski limit".
So any assumption about the association rate constant could easily be out
by a factor of 10^4, which in turn gives you a 10^4 error in calculating
the koff from the Kd!
Steve Halford
----------------------
Professor Stephen E. Halford FRS
Department of Biochemistry,
School Of Medical Sciences,
University of Bristol,
University Walk,
Bristol BS8 1TD, UK.
tel: +44-(0)117-331-2156 - NB, new phone and new FAX number
FAX: +44-(0)117-331-2168
E-mail: s.halford at bristol.ac.uk
--On 07 April 2008 08:20 -0400 Allen Minton <minton at helix.nih.gov> wrote:
> If memory serves, many if not most protein association reactions have
> very similar bimolecular association rate constants (say within a factor
> of 3 or so). I seem to remember something on the order of 10^5 mol^-1
> sec^-1, but I'm not sure of that, you'd better look it up. (Note: this
> is orders of magnitude smaller than the upper limit calculated according
> to Smoluchowski, which does not take into account the orientational
> constraints applying to a specific association that greatly increase the
> entropic barrier to complex formation.) Therefore, given the equilibrium
> dissociation constant you should be able to estimate the dissociation
> rate constant to within this level of uncertainty.
>
> Diffusion coefficients are irrelevant.
>
> Allen Minton
>
>
> At 07:59 PM 4/5/2008, Wenta Nikola wrote:
>
> Content-class: urn:content-classes:message
> Content-Type: multipart/alternative;
> boundary="----_=_NextPart_001_01C89779.0B736A53"
>
> Dear Colleagues!
> In order to make my question more precise, I want to to calculate the
> theoretically diffusion-controlled maximum possible k_off rate of a given
> dimer based it's kD (from SE run), in combination with a calculated
> diffusion coefficient for monomer and dimer by HydroPro.
> If anyone had any suggestions how to calculate that k_off rate, I would
> be very grateful.
> Niko
>
> This message has been checked for viruses but the contents of an
> attachment may still contain software viruses, which could damage your
> computer system: you are advised to perform your own checks. Email
> communications with the University of Nottingham may be monitored as
> permitted by UK legislation.
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
>
More information about the RASMB