[RASMB] Lamp Diagnostics
John Gunther
jrg at mywdo.com
Sat Apr 5 06:34:48 PDT 2008
Borris,
I believe it's because there is a windowless cell. Often I run the holimum oxide cell which has a 3krpm speed limit.
Then I do a wavelength scan on the holimum oxide cell.
Also you would have to wait for a 50u vacuum, and wait until it gets to 60k to do the test.
You can do the 3k test without a 50u vacuum. It takes only a few minutes.
So actually you get done faster @3k.
John Gunther
*********** REPLY SEPARATOR ***********
On 4/5/2008 at 7:48 AM Borries Demeler wrote:
>> Content-Transfer-Encoding: quoted-printable
>>
>> Here is the lamp intensity check procedure from a 10-2001 manual. I'd=
>> follow this to step 8 then just look at it. The origin part is a waste
>of=
>> time in my opinion.
>> John Gunther
>
>John,
>
>do you happen to know what the significance of running this test at 3000
>rpm is?
>Why not at 60 k, where you get done faster?
>
>-Borries
>
>
>--
>No virus found in this incoming message.
>Checked by AVG.
>Version: 7.5.519 / Virus Database: 269.22.7/1361 - Release Date: 4/5/2008
>7:53 AM