[RASMB] Protein degradation in AUC
Leech, AP
apl3 at york.ac.uk
Fri Oct 12 05:34:21 PDT 2007
Hi Claus, all,
I have given up with pipe-cleaners as they usually have a wire down the
middle which risks scratching. I now use Chemtronics #140 Foamtips which
are small enough to fit into the sectors but all plastic construction.
These seem to be good for removing mechanically adherent residues that
liquid reagents don't dissolve.
I usually keep the residue from sample loading in the lab at about the
same temperature as the run, so if necessary the user can run a
comparison gel after the experiment - sometimes the protease is in the
sample before it gets to the cell, and the time at 20C is the problem!
Andrew
Claus Urbanke wrote:
> Hello RASMBers,
>
> for a few weeks now we encounter a nasty problem regarding the cleaning
> of epon centerpieces (charcoal/aluminum). With an occurence of 10 - 30%
> protein will be partially cleaved (by a protease?) after a sedimentation
> run. Has anyone encountered this before? How can one clean the
> centerpieces. The resistance chart tells me that GuaHCl is satisfactory
> and urea is unknown, so we are testing some centerpiece fragments over
> the weekend in concentrated urea solution.
>
> We are cleaning our centerpieces with a pipe cleaner using commercial
> household detergent rinse them with hot water and finally with doubly
> distilled (dd) water. The detergent itself was tested to be protease
> free. We use saphire windows that are cleaned by > 1 hour 50% nitric
> acid, rinsing with water, dd water and drying.
>
> Many thanks in advance for any suggestion
>
> Claus
>
>
>
--