[RASMB] Multiple wavelength data
Virgil
schirfv at biochem.uthscsa.edu
Wed Apr 11 06:15:55 PDT 2007
You would have to run it parallel to the stem, otherwise you will cut
off the 280 nm light.
Cheers,
Virgil
Gu-Gang Chang wrote:
> Hi
>
> I have a simple technical question. Can I collect UV (<400 nm) and Visible
> (>400 nm) data simultaneously for the same sample? There is a blocking
> filter in the XL-A monochromator. If I want collect AUC data at two
> wavelengths 280 nm and 410 nm, which position should I choose? Parallel or
> perpendicular to stem?
>
> Thanks
> ggc
>
>