[RASMB] measuring refractive index in the near IR, for light scattering apps.
Richard Kingston
rl.kingston at auckland.ac.nz
Sun Feb 4 19:25:27 PST 2007
Greetings all,
I am seeking advice about measuring the refractive index of a
solution, which crops up when analyzing light scattering data.
Most who are familiar with the theory of dynamic light scattering
will know that the relationship between the measured auto-
correlation function and the diffusion constant involves the
refractive index of the medium. To get a real number for the
diffusion constant of a protein from DLS, you need the refractive
index. Yet the laser light sources used on commercial instruments
(from Wyatt or the now assimilated DynaPro) often operate in the near
IR (e.g. 830-840 nm). Measuring the refractive index here is non-
trivial. Most lab refractometers are calibrated to give the
refractive index at 589 nm (The Sodium D line).
The options seem to be ... in order of increasing correctness and
decreasing ease.
1. Pretend your solution is water, and look the numbers up in a table.
2. Measure the refractive index of the solution using a standard Abbe
refractometer and ignore the dispersion (wavelength-dependence).
3. Directly measure the refractive index at the wavelength in question.
Option 1 starts looking bad in some circumstances (e.g. when adding
salt or osmolyte to the solution)
Option 2 is not difficult, but I'm not certain how large an error is
introduced by ignoring the wavelength dependence of the refractive
index (say between 589 and 840 nM). Looking at data for pure solvents
(water, ethanol, glycerol) it seems like the error introduced could
be small.
Option 3 is made difficult by a lack of cheap, accessible
instrumentation. Recently I came across an interesting idea which
would allow measurement of refractive index of solutions in the near
IR without the need for expensive equipment. The reference is ...
J Rheims, J Köser and T Wriedt (1997) Refractive-index measurements
in the near-IR using an Abbe refractometer. Meas. Sci. Technol. 8
601-605
It's easy to find the full text of the article using Google. The
basic concept is to replace the light source on a conventional Abbe
refractometer with an IR laser, and attach an IR sensitive CCD to the
refractometer eyepiece tube. Then the instrument must be re-
calibrated to make sense of the readings.
This is where it becomes problematic for owners of old Bausch and
Lomb Abbe-3Ls (surely the most common refractometer ever produced ?)
The re-calibration requires knowledge of the glass used in the
refracting prism. Yet generally, all you have to go on is the prism
series number - engraved on the prism, and on the left of the
instrument scale. In my case it reads "538". Although I have a copy
of the Bausch and Lomb glass catalog, the company long ago got
swallowed by others, and I don't know a way of telling what glass was
used in a refracting prism of any given series. That's likely true
for most of the Bausch and Lomb Abbe-3Ls except those in current
manufacture. So this route is blocked.
There doesn't seem to be a lot of discussion of the general problem
in the literature, which might indicate that measuring the refractive
index in the near IR is unimportant. Or it might not. I'd appreciate
any comments from light scattering specialists or others who might
have thought about this issue.
Thanks,
Richard
Richard Kingston, PhD.
School of Biological Sciences
The University of Auckland
Private Bag 92019
Auckland
New Zealand.