[RASMB] Vacuum pump

John Philo jphilo at mailway.com
Sun Jun 25 09:27:02 PDT 2006


Christine,

I am wondering if your real problem is a cooling problem rather than a pump
problem. Have your service engineer check that all the cooling fans are
running properly, and that the air outlets are not clogged with dust.

John

-----Original Message-----
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Christine.Wandrey at epfl.ch
Sent: Sunday, June 25, 2006 2:34 AM
To: rasmb at rasmb.bbri.org
Subject: [RASMB] Vacuum pump


Dear all,

two weeks ago, after more than 8 years, the 
vacuum pump in our  XL-I died (the pump became 
very noisy).
The Beckman service first changed the motor. 
Though the first run after this reparation 
stopped after about 4 hours with the message that 
there was no vacuum after running about 4 h at 40 
000 rpm.
Then the service recommended to replace the old 
pump by a new one. We agreed and the service 
arrived with a new one. This one became very hot 
already during the test run. The service had to 
come again with a another new pump. With the new 
pump the 0 microns were reached after 20 nim. 
However, after the departure of the service, the 
subsequent run stopped after about 2 hours with 
the same message: bad vacuum.
The service came again now running the XL-I and 
measuring the temperature in the pump. After 
about 4 hours the pump/oil temperature approached 
75°C and stopped, with the consequence that the 
AUC stopped running too.
Having identified the problem there is my 
questing:  Is there a special pump type that I 
should request from the service in order to avoid 
this problem? Has anybody similar experience?

Thank you in advance for your recommendations.

With kind regards,
Christine
-- 
PD Dr.habil. Christine Wandrey, MER
Ecole Polytechnique Fédérale de Lausanne (EPFL) EPFL-SB-ISIC-LMRP2, station
15 CH-1015 Lausanne Suisse

tél: (+41) (21) 693 36 72 in CH J2 494, Station 6
  or (+41) (21) 693  96 61 in AAB 042, Station 15
fax: (+41) (21) 693 60 30

_______________________________________________
RASMB mailing list
RASMB at rasmb.bbri.org http://rasmb.bbri.org/mailman/listinfo/rasmb





More information about the RASMB mailing list