Fri May 12 11:09:54 PDT 2006


Dear all,

I have done experiments at 30C using XL-I and it worked superbly.

Regards,

Trushar

**********************************************
Trushar Patel
PhD Student
National Centre for Macromolecular Hydrodynamics
School of Biosciences
University of Nottingham
Sutton Bonington
LE12 5RD, UK
Tel: +44 115 9516149(Lab)

E-mail : stxtrp at nottingham.ac.uk
www: http://www.nottingham.ac.uk/ncmh/students/trushar.html
***********************************************
>>> "Christine Ebel" <christine.ebel at ibs.fr> 05/12/06 6:37 PM >>>
Hello,

I have a bad experience with my unique run at 37°C.

I had oil condensation.

In consequence, for this run, and all the experiments after, (until the
Beckman man came and washed everything) even if I washed the filter, where
consensation was obvious, interference was unusable: it become noisy (very
noisy) after about two hours of run.

The problem is not so obvious with absorbance.

I am curious to know at what temperature we can run using interference
optics.

Christine

 

Christine EBEL

Institut de Biologie Structurale UMR 5075 CEA-CNRS-UJF

41 rue Jules Horowitz, 38027 Grenoble Cedex 1, France

Tel (33) (0) 4 38 78 96 38; Fax (33) (0) 4 38 78 54 94 

christine.ebel at ibs.fr

 <http://www.ibs.fr/content/ibs_eng/presentation/lab/lbm/ebel.htm>
http://www.ibs.fr/content/ibs_eng/presentation/lab/lbm/ebel.htm

-----Message d'origine-----
De : rasmb-bounces at server1.bbri.org [mailto:rasmb-bounces at server1.bbri.org]
De la part de Dean Malencik
Envoyé : vendredi 12 mai 2006 18:28
À : RASMB at server1.bbri.org
Objet : [RASMB] XLA run at 37 degrees

 

Hello,

 

I have been asked to do a run at 37 degrees on the XLA analytical
centrifuge.  Does anyone have any pitfalls to anticipate or is it even
feasible to do a run at this temperature?  Any differences if the run is a
velocity or equilibrium run? 

 

Thanks.

 

Dean Malencik



This message has been checked for viruses but the contents of an attachment
may still contain software viruses, which could damage your computer system:
you are advised to perform your own checks. Email communications with the
University of Nottingham may be monitored as permitted by UK legislation.




More information about the RASMB mailing list