[RASMB] XLA run at 37 degrees
Christine Ebel
christine.ebel at ibs.fr
Fri May 12 10:37:05 PDT 2006
Hello,
I have a bad experience with my unique run at 37°C.
I had oil condensation.
In consequence, for this run, and all the experiments after, (until the
Beckman man came and washed everything) even if I washed the filter, where
consensation was obvious, interference was unusable: it become noisy (very
noisy) after about two hours of run.
The problem is not so obvious with absorbance.
I am curious to know at what temperature we can run using interference
optics.
Christine
Christine EBEL
Institut de Biologie Structurale UMR 5075 CEA-CNRS-UJF
41 rue Jules Horowitz, 38027 Grenoble Cedex 1, France
Tel (33) (0) 4 38 78 96 38; Fax (33) (0) 4 38 78 54 94
christine.ebel at ibs.fr
<http://www.ibs.fr/content/ibs_eng/presentation/lab/lbm/ebel.htm>
http://www.ibs.fr/content/ibs_eng/presentation/lab/lbm/ebel.htm
-----Message d'origine-----
De : rasmb-bounces at server1.bbri.org [mailto:rasmb-bounces at server1.bbri.org]
De la part de Dean Malencik
Envoyé : vendredi 12 mai 2006 18:28
À : RASMB at server1.bbri.org
Objet : [RASMB] XLA run at 37 degrees
Hello,
I have been asked to do a run at 37 degrees on the XLA analytical
centrifuge. Does anyone have any pitfalls to anticipate or is it even
feasible to do a run at this temperature? Any differences if the run is a
velocity or equilibrium run?
Thanks.
Dean Malencik
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20060512/712a876a/attachment.htm>
More information about the RASMB
mailing list