Norbert,

Yes, Andy is right, this is due to deflection of the light beam by the
refractive index gradient caused by the boundary itself. The 3 mm pathlength
cells should allow you to go to 6 mg/ml, although you still will have issues
with data analysis due to the significant non-ideality effects.

John

-----Original Message-----
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Norbert Muecke
Sent: Friday, October 28, 2005 4:29 AM
To: RASMB at server1.bbri.org
Subject: [RASMB] peak in sedimentation velocity scans at high protein conc.


Hi,
I have done some velocity runs at high protein conc. (up to 6g/l). A new
peak within the boundary could be observed. I have attached the scans of
three different concentrations. Absorbance was measured at 284nm. Any idea
of how to explain that phenomena would be appreciated.

Best regards
Norbert