I seriously doubt that you will find a much better way to calibrate
temperature than the CoCl2 approach Walter published. Trying to use
solubility introduces factors like concentration & nucleation effects
into the mix. It also only allows a measurement at one temperature.
At the intro & service Beckman courses in Palo Alto at the end of one
of the days they half-fill two sectors of a prep rotor with vacuum or
drive oil, spin overnight, & put a thermometer into the solutions to
check temperature. Its one point per day but I guess a good way to
check even for prep centrifuges.
I certainly think this standardization is a good idea, & I am glad to
participate, although I must stress again I calibrate temperature and we
measure density, and density*vbar is a much larger source of error. So
get a density meter while you're at it!
I would also encourage you to recognize this is not the experts coming
to the rescue. Any physical measurement requires calibration and
careful regulation of temperature. This is a fact of life, not a
special event that will somehow resolve the problem for everybody & all
machines. You must wait for temperature before hitting start, & if you
want to do careful work you must calibrate. (I asked for a poll earlier
- if you have calibrated temperature on your XLA/I send me a note!)
The suggestions to do replicates is not something unique to this event.
You must always do replicates, as a function of concentration, to
understand the behavior of your system. The sample Arthur sends to us
will presumably be nonassociating, low charge, homogeneous, & in a
buffer near the isoelectric point. For typical samples association,
heterogeneity or nonideal behavior are the sources of deviation. In my
lab multiple users typically achieve 0.1 - 0.3% reproducible S values on
well behaved noninteracting systems, & this applies to joint analysis
with sedfit, DCDT+ and SEDANAL. (For monomers the analysis method
should not be that critical - but it is the holidays, so get in the
spirit & sell away.) I also honestly think users doing their own
analysis should be part of the experiment, for no reason other than
setting a standard & expecting the field to meet that standard. I think
I have now crossed the PC line -
happy holidays
-------------------------------------------------------------------
Dr. John J. "Jack" Correia
Department of Biochemistry
University of Mississippi Medical Center
2500 North State Street
Jackson, MS 39216
(601) 984-1522
fax (601) 984-1501
email address: jcorreia at biochem.umsmed.edu
homepage location: http://biochemistry.umc.edu/correia.html
dept homepage location: http://biochemistry.umc.edu/
-------------------------------------------------------------------
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20041222/0f7b4585/attachment-0001.html>
More information about the RASMB
mailing list