[RASMB] difference of p = 0.95, 0.68 and 0.55, the confidence level in the sedfit c(s) distribution
Arthur Rowe
arthur.rowe at nottingham.ac.uk
Mon Jul 12 12:18:01 PDT 2004
Hi all
In Peter's absence let me make one point which he (very correctly) makes
concerning estimation of s values and relative concentrations. Which is that
the c(s) distribution for each individual species in a mixture is not a
gaussian distribution. Hence one should not do 'peak fitting' by the usual
algorithms, in ORIGIN or anything else. As Jack very correctly says, it is
numerical integration over appropriate ranges which you need.
What is valid to a very good approximation is to fit multiple gaussians to
the (least squares in SEDFIT) g(s) profile. Basically as per Walter
Stafford's original approach. In our own experience we find this gives a
more objective description of such systems, albeit - since it is better at
resolving closely related species including dimers of lower M (i.e. more
rapidly diffusing) monomers - the outcome can be rather less flattering to
the perceived 'quality' of one's precious preparation!
All best wishes to everyone
Arthur
--
*************************
Arthur Rowe
Lab at Sutton Bonington
tel: +44 115 951 6156
fax: +44 115 951 6157
*************************
Since Peter is on a long vacation, I will attempt to answer your question.
You can integrate the peaks by pressing the ctrl and I keys simultaneously
which will give you a dialog box that states that you should hold the right
mouse button down and draw a rectangle to cover the s range to be
integrated. You will see once you do the integration you will obtain both
the % of the loading signal in the integration range and the weight average
s value. Also the results box states that this integration it is best done
without regularization, confidence level of zero. Regularization gives the
most parsimonious distribution for the confidence level that you set. Total
removal of regularization may give you too many peaks that will merge at
higher confidence levels. You can still integrate over multiple peaks in the
s range you have selected and compare the result with integration of the
distribution you obtain at higher confidence levels. In my experience the
integration results over the same s range are comparable from no
regularization to using settings of p = 0.68 to 0.95. At the latter p
selections you have the more realistic description of the sedimenting
species. Hope that the above is clear.
Jack Lebowitz
At 10:29 PM 7/12/2004 +0800, medakachou wrote:
Dear all,
Recently, I'm analyzing the sedimentation velocity spectra by continuous
c(s) distribution (SEDFIT). I've analyzed my data in three kind of
confidence level: p = 0.95, 0.68 and 0.55 and the regularization method is
maximum entropy. The s limit is 0.1 to 25S. I found every species is not
well seperated (they just fuse together) in p = 0.95. In p = 0.68, the
situation is better and the peaks are more significant. p= 0.55 can give me
the highest resolution and every peak is very clear cut. Now the question
is: if I want to calculate the area of peaks by Origin peak fitting module,
which results should I use? I've check Schuck's paper and he suggests using
p = 0.68 to 0.95 is enough. How about 0.55? I appreciate your response and
suggestion.
Sincerely,
Chi-Yuan Chou
PhD student, the Institutes of Life sciences, National Defense Medical
Center, Taipei, Taiwan
e-mail: r6243023 at yahoo.com.tw
This message has been scanned but we cannot guarantee that it and any
attachments are free from viruses or other damaging content: you are
advised to perform your own checks. Email communications with the
University of Nottingham may be monitored as permitted by UK legislation.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20040712/1fe9c53e/attachment.html>
More information about the RASMB
mailing list