[RASMB] back to reality ... and overspeeding
Walter Stafford
stafford at bbri.org
Fri Oct 1 08:26:00 PDT 2004
All:
Ariel's point is well taken. It reminded me of
two other studies on overspeeding:
One very detailed theoretical study from David
teller's group in the 1960's. (cf page 78: D.C.
Teller and T.A. Horbett, "Ultracentrifuge Studies
with Rayleigh Interference Optics. III ...",
Annals NY Academy Sciences pp 66-101.) gives over
speeding protocols. It also warns of the serious
dangers of over - overspeeding. In 1976 Dennis
Roark published a handy little formula for
calculating overspeed times based on the
calculations done by the Teller Lab. The
overspeeding procedures cited can - if properly
applied (and properly is the key word here) -
reduce the time to equilibrium to about 1/6 th
the time not using overspeeding. However, if you
look at the figure on page 78, you'll see that
the choice of overspeeding time is quite
critical. Use of overspeeding with unknown mixed
systems could lead to undesirable results since
each species has a very narrow useable
overspeeding time.
Here is the reference (and abstract) to Denny's paper.
Roark, D. E. (1976). "Sedimentation equilibrium
techniques: multiple speed analyses and an
overspeed procedure." Biophysical Chemistry
5(1-2): 185-96.
The use of a single high-speed
sedimentation equilibrium experiment to analyze
mixed associating systems is inadequate to
determine the association mode even if the
molecular weight of one species is known.
Simultaneous analysis of the concentration
distributions at three equilibrium speeds greatly
reduces the lack of uniqueness. Linear
least-squares multi-speed fits discriminate
between association models in which the molecular
weights are assumed. Experiments at a series of
initial concentrations as well as rotor speeds
further increases the discrimination. An
overspeed procedure is proposed. The overspeed
time depends only slightly on the sedimenting
species molecular weight, but significantly on
the frictional ratio. A minimal overspeed time
may be estimated.
Walter
p.s. Also a must read from David Teller's Lab:
Teller, DC, Characterization of proteins by
sedimentation equilibrium in the analytical
ultracentrifuge. Methods Enzymol. 1973;27:346-441.
At 5:58 PM +0200 9/29/04, Ariel Lustig wrote:
>Dear colleauges, lets come back to reality
>At least for new-comers , but not only for
>them, the subject of predicting or better ,
>calculating the time to achieve equilibrium
>makes no / or only very little sense , since as
>I belive that every one of us use the
>"overspeed procedure" discribed by Hexner,
>Radford and Beams PNAS (1961), 47,1848 that
>reduces time to achieve equilibrium to 50% or
>more. If so, I ask myself what means time to
>equilibrium ? from the time rotor speed is
>achieved and not yet a concentration gradient is
>established, or a steep graient at both
>solution column ends? all non -sense ! I agree
>that for polydispersed or selfassociating
>systems
>using an overspeed procedure is difficult (and
>dangrous especially it some parts are not
>reversible), but not less difficult as to
>calculate time to equilibrium as you suggested
>of a non monodispersed systeme.
>In FRACTION N° 1 1967 issued by Beckman
>Instruments ( may be based on the same equation
>that Walter Stafford mentioned recently) K.E
>Van Holde discribed a formula where the time
>depends on Diffusion ! and here again to
>new-comers. If at all making the exercise to
>calculate time to equilibrium.
>It is to use Dobs and not D20w that is
>correected for Viscosity of water. There may
>be a huge difference
>between the two Diffusion values if per
>example you use sucrose in the buffer as I do
>measuring
>membrane proteins. The density has no influence
>for Diffusion only the viscosity of the
>solvent.
>yours....ariel
>ariel.lustig at bluewin.ch
--
----------------------------------------------------------
Walter Stafford
mailto:stafford at bbri.org
direct dial: 617-658-7808
receptionist: 617-658-7700
----------------------------------------------------------
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20041001/1b370b22/attachment.htm>
More information about the RASMB
mailing list