[RASMB] difference between experimental and modelled sed coef for protein-DNA complexes
Peter Schuck
pschuck at helix.nih.gov
Thu Nov 20 17:35:00 PST 2003
Marcelo,
Dependent on the rotor speed you've used, I also think it is possible that
the glycerol will sediment and form density and viscosity gradient. I
would do an experiment like Borries suggested and run the glycerol/D2O
mixture against water in the XLI. If you see a gradient forming, it likely
will significantly influence your macromolecular sedimentation. I've
recently examined such effects in theory. You can find a description of
the results and a SEDFIT model how to quantitatively take this into account
when modeling at
http://www.analyticalultracentrifugation.com/inhomogeneous_solvent.htm
Another problem could be that at the high concentration of glycerol you
might run into effects of preferential hydration leading to deviations of
your v-bar from the calculated one.
Best,
Peter
More information about the RASMB
mailing list