Borris (and anyone else),
For a global analysis of sedimentation equilibrium data obtained over a range of wavelengths, do you use extinction coefficients calculated using the XLA absorbance optics, or do you use values obtained with more precise insturmentation? Often, when one sets a scan at 230nm, the value written to the data file is off +/- 1nm. At 230nm, this can change the absorbance by a significant amount. Which value is correct - the one entered or the one written to the data file? Finally, you indicated you don't trust data below 0.1 OD. If you have collected an absorbance scan that ranges from below 0.1 to a value around, say, 0.9, do you exclude data below 0.1 in the analysis? I agree one should not trust data if the entire scan is below 0.1 OD, but if the data spans a wide range of OD's, what is the most appropriate data to include?
Thanks,
N. Karl Maluf
-----Original Message-----
From: rasmb-admin at rasmb-email.bbri.org on behalf of Borries Demeler
Sent: Tue 12/16/2003 7:33 PM
To: Lin Fang
Cc: rasmb at alpha.bbri.org
Subject: Re: [RASMB] Loading concentration
----------------------------------------------------------------------------------
The older archived RASMB emails can be found at:
http://rasmb-email.bbri.org/rasmb_archives
and current archives at
http://rasmb-email.bbri.org/pipermail/rasmb/
Search All the Archives at:
http://rasmb-email.bbri.org/rasmb_search.html
----------------------------------------------------------------------------------
>
> Hi,
>
> I got my data for equilibrium analytical ultracentrifugation. It
> turned out the absorbance of my protein is below 0.1 (OD280). Are
> these data still analyzable? And what is the optimum range for
> equilibrium analytical ultracentrifugation? Thanks a lot.
>
>
> Lin
Lin,
you can probably shift the wavelength to 230 nm or lower to see if
your protein absorbs better. You need to make sure that your buffer
doesn't absorb at the lower wavelength. I generally include absorbance
data up to 0.9 OD, give or take a few dependening on wavelength intensity.
less than 0.1 OD sounds almost like nothing. At this OD 10% of the signal
is noise which is too much to get reliable analysis for any system.
It is best to use multiple concentration measurements and include all
of them in your global fit. I generally use 2 6-channel cells with
0.3, 0.5 and 0.7 OD 230 and 280. This gives me a broad concentration
range over which I am analyzing the sample. For many cases this works
very well. If you don't have aromatic sidechains in your protein, your
mileage may vary at 280 nm.
Good luck, -Borries
_______________________________________________
RASMB mailing list
RASMB at rasmb-email.bbri.org
http://rasmb-email.bbri.org/mailman/listinfo/rasmb
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20031216/eaa4968e/attachment.html>
More information about the RASMB
mailing list