[RASMB] Can DCDT+ be used to detect small change in protein conformation? What is its limit?
Walter Stafford
stafford at bbri.org
Sun Sep 22 19:58:00 PDT 2002
Hi Chris,
Yes, DCDT can be used to detect very small changes in s. If
you match the menisi using a capillary type synthetic boundary cell
so the the sample on each side starts sedimenting at the same
position. You can iether compare two samples directlyIf you are
adding a low molecular weight ligand, you must put the sample plus
ligand on the "sample" side , fill the "reference" side slightly more
(20uL) than the sample side with sample without ligand. Take the run
up to 3-10K rpm to match the menisci. Then stop the run, gently shake
the rotor to obtain a uniform distribution and restart. A small
amount of sample containing no ligand (10uL) will flow in to the
sample and cause a very small dilution of ligand. The sample must
have been dialyzed vs. the basic buffer without ligand before being
divided. When you add ligand to the sample also add an equal amount
of buffer to sample not containing buffer to dilute it by the same
amount. The ligand should be dissolved in the dialysate. Am assuming
that you are using the interference optics. The data can then be
analyzed with DCDT. by fitting the boundary to two gaussians - on
positive and one negative. I will give all the details if you contact
me directly. You can also fit the data using numerical solutions to
the Lamm equation using SedAnal.
Walter Stafford
At 13:55 -0500 9/20/02, Chin, Christopher wrote:
>To John, all DCDT+ users and all interested colleagues,
>
> John, I would like to have your comment in the following three
>experiments using RNAseA as a model and DCDT+ as the method to do
>data analysis.
>
>Experiment 1: The purpose is to check the stability of RNAseA. The
>comparison is made between the protein left at room temperature
>for a couple of days and that storage in the freezer.
>Experiment 2: Question: Can the denatured RNaseA (has 4 s-s bond)
>in GuHCl be refolded back to its native state in the absence of
>reducing agent?
>Experiment 3: Question: Can the denatured RNaseA (has 4 s-s bond)
>in GuHCl be refolded back to its native state in the presence of
>reducing agent TBP (Tri-n-butylphosphine).
>
>The major issue here is how I can differentiate the DS difference
>due to scan chosen from the DS difference due to real experimental
>finding (between the control and the sample).
>
>I am pleased to say that, using the criteria that I have stated,
>e.g. maximized or magnified the DS difference due to scan
>chosen (use a smaller number for denominator when converted to
>%), minimized or play down the DS difference between the control and
>the sample (use a larger number for denominator when converted
>to % ), the results are pretty much I have expected. Do you think
>I try to push the instrument limit too far to obtain these results
>by manipulating the treatment of data?
>
>s20,w values reported here are the results of fitting dcdt.
>
>Thank you in advance for your comment or advice.
>
>Chris
>
>
>--------------------------------------------------------------------------
>Christopher Chin
>Manager, XLA-Analytical Ultracentrifugation facility
>Sealy Center for Structural Biology
>HBC&G, 5.134 MRB.UTMB, Galveston,TX 77555-1055
>cchin at utmb.edu, 409-772-1693, efax 708-585-1920
>---------------------------------------------------------------------------
>
>
>
>Attachment converted: WFS3HD:Stability of RNAse.doc (WDBN/MSWD) (00121490)
>Attachment converted: WFS3HD:GuHCl treated(without reducing
>(WDBN/MSWD) (00121491)
>Attachment converted: WFS3HD:GuHCl-TBP treated and renatural
>(WDBN/MSWD) (00121492)
--
########################################################################
Walter F. Stafford III, Ph.D.
Senior Scientist
Analytical Ultracentrifugation Research Laboratory
Boston Biomedical Research Institute
64 Grove Street
Watertown, MA 02472-2829
and
Associate in Neurology
Massachusetts General Hospital
Harvard Medical School
Boston MA 02115
main:(617) 926-8040
tel: (617) 658-7808
fax: (617) 972-1753
mailto:stafford at bbri.org
http://www.bbri.org/faculty/stafford/Stafford.html
#################################################################
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20020922/f25dcae4/attachment-0001.html>
More information about the RASMB
mailing list