Dear Dear RASMB colleagues ,
Dear Songpon,
Question 1. I believe , if you talk about S values you mean S° w,20 (I
mean S extrapolated to
Zero concentration) .Alongated rod -stractur are very concentration
dependent, but we have to be careful with extrapolation to zero conc.,
often at low conc. the particle may dissociate in
subunits so this extrapolation makes no sense.To proof this, a row of SE
runs at various conc. would help to understand the system.
Diffusion runs are hard to trust in ! The best one are those done
with schlieren optics at high conc. and ideal in behavior .
Non-symmetrical and conc. Dependent once includes the diffusion
coefficient of more then the one conc. Diffusion runs
done with absorption optics have rarely
a good baseline (to indicate precisely the net area of such an curve)
what influences mainly the value.
Diff. calculated from s' and M` are not accurate but possible to use.
My opinion is that 7.2 and 6.8 S° 20w !! are not too far from each other.
If M from SE can be
achieved dont worry about the diffusion.
Questio 2.
My detergent experience is not with LDAO, but only with Octyl Poe,
DDM and OG.
It is worth to read the references and the problematic use of detergents!!
A.Lustig, A.Engel, G.Tsiotis ,E.M. Landau , W.Baschong. Molecular weight
determination of membrane proteins by sedimentation equilibrium at the
sucrose or Nycodenzadjusted density of the hydrated detergent micelle,
Biochim. Biophys. Acta 1464 (2000)199-206
J.P. Rosenbusch , A. Lustig , M. Grabo, M. Zulauf and M.Regenass,
Approaches to determining membrane protein structures to high resolution:
do selections of subpopulation occur? Micron 32, (2001)75-90
Let say the detergent run is not the problem only the ATP, so if possible
why not to make such a run with the XLI or model E with interference ,
there the ATP has no influence.
I myself used ATP solutions with the XLA but scanned at 285- 288
nm.there the ATP
has minor influence. This is possible if some part of your prot
absorbs at 280nm.
If not only the refraction ....optics can help.
What concerns the baseline, may be the SEGAL computer program may help
here is the address
http://www.biozentrum.unibas.ch/personal/jseelig/AUC/index.html
with the intention to help.......yours.........ariel